rietveld_l  

RE: Negative Biso / absorption in TOPAS.

Nicolich, Jeffrey
Wed, 10 Feb 2010 06:55:31 -0800

 Is it possible the data was collected with variable slit aperture? You
could try to convert the data to fixed slit aperture intensities and see
what happens to the B values.

Regards,
Jeff Nicolich

W.R. Grace & Co. - Conn.
62 Whittemore Ave, Cambridge, MA 02140
+1-617-498-3816
jeffrey.nicol...@grace.com

-----Original Message-----
From: Alan Hewat [mailto:he...@ill.fr] 
Sent: Wednesday, February 10, 2010 5:43 AM
To: Jacco van de Streek
Cc: rietveld_l@ill.fr; alan.he...@neutronoptics.com
Subject: Re: Negative Biso / absorption in TOPAS.

> 1. Is absorption indeed the most likely cause of this problem or are 
> there other possibilities?
>
> 2. How do we model absorption in TOPAS? Which keyword do we use and 
> what are reasonable values for the parameters in the expression?

1. You will find discussions of negative B-factors in the Rietveld list
archive http://www.mail-archive.com/rietveld_l@ill.fr/info.html Possible
causes cited for X-rays include polarisation correction, surface
roughness and absorption. You really do need to know how the data were
collected - Rietveld refinement should not be treated as a black box.

2. If B-factors are negative, it obviously means that higher order
reflections are stronger than they should be. Absorption can cause this,
since absorption will be greater for transmission (low angles). For a
cylindrical sample it can be shown that moderate absorption has exactly
the same effect on intensities as a negative contribution to the overall
B factor, and indeed you can calculate the magnitude of the effect if
you know the absorption coefficient. Again, use physical information
rather than simply refining more parameters.
_____________________________________________
Dr Alan Hewat, NeutronOptics, Grenoble, FRANCE
<alan.he...@neutronoptics.com> +33.476.98.41.68
      http://www.NeutronOptics.com/hewat
______________________________________________