Pulmonetic Systems LTV Series of Ventilators
Audience: Respiratory
healthcare professionals and risk managers
FDA and Pulmonetic Systems, Inc. notified healthcare
professionals of a Class I recall of the LTV series of ventilators,
models 1000, 950, 900 and 800, designed to automatically switch to
internal battery operation, allowing uninterrupted ventilation, when an
external power source is removed or is no longer adequate to power the
ventilator. The ventilators malfunction when switching to the internal
battery, causing failure of the ventilator to breathe for the patient.
http://www.fda.gov/cdrh/recalls/recall-090804.html
The following document is a copy of a Press Release from
Pulmonetic Systems, Inc. releated to the
Class I
Medical Device Recall of the Pulmonetic Systems LTV Series of
Ventilators.
FOR IMMEDIATE RELEASE
Monday, September 27, 2004
FOR: Pulmonetic Systems, Inc.
17400 Medina Road, Suite 100
Minneapolis, MN 55447
CONTACT: Dave Neuger
Neuger Communications Group
507.664.0704 (direct)
888.761.3400 (toll free)
[EMAIL PROTECTED]
PULMONETIC SYSTEMS, INC. ISSUES
NATIONWIDE RECALL
OF CERTAIN LTV VENTILATORS
MINNEAPOLIS – Pulmonetic Systems, Inc., Minneapolis,
Minnesota, is voluntarily initiating a recall of approximately 10,000 of
its portable ventilators intended for mechanical ventilation of adults
and children in homecare, hospitals and other institutions. The
ventilators being recalled were manufactured before September 1,
2003.
The ventilators malfunction when switching to the internal battery when
the external power source is inadequate. This problem causes failure of
the ventilator to breathe for the patient, which could result in
permanent neurologic injury or death.
The company has received approximately 30 reports of this malfunction in
which injury was averted by prompt caregiver action. The company has
received one report of a patient death possibly associated with this
situation which remains under investigation.
A Safety Alert, accompanied by a Caregiver Information Sheet, was sent
out by the company as an interim solution on September 7, 2004, to inform
customers and users of this potential malfunction. These publications
describe the malfunction, the alarms that should occur, and give detailed
instructions to users on how to immediately correct the malfunction and
resume appropriate ventilation. If instructions in the Safety Alert are
properly followed, the unit will usually continue to provide
ventilation.
There is the remote possibility that when switching to internal battery
operation, the ventilator may not operate after disconnecting from an
inadequate external power supply, if the internal battery is not properly
charged. External DC power sources include three and nine hour lead-acid
batteries as well as automobile batteries powering the ventilator through
a cigarette lighter adapter. As identified in the Safety Alert, and as
part of ventilator patient care, caregivers should always be prepared to
provide alternate means of ventilation should any product malfunction
occur.
The company expects to provide a permanent solution for all affected
devices that can be easily installed by the caregiver beginning within
the next 4-6 weeks. Meanwhile, users and caregivers need to read the
Caregiver Information Sheet to understand how to react in the event this
malfunction should occur.
The recall includes all LTV series ventilators manufactured before
September 1, 2003 unless they are specifically marked on the back panel
label of the ventilator with a symbol of a music note in the upper left
or lower right hand corners. The following models and serial numbers may
be affected:
LTV 1000 Serial Numbers below A06500
LTV 950 Serial Numbers below C07000
LTV 900 Serial Numbers below B03500
LTV 800 Serial Numbers below D01400
The firm voluntarily took this action and the FDA has been apprised. This
action has been classified by the Food and Drug Administration (FDA) as a
Class I Recall. The FDA defines a Class I recall as a situation in which
there is a reasonable probability that the use of or exposure to the
product will cause serious adverse health consequences or
death.
Pulmonetic Systems has notified affected device owners by certified mail.
Users with affected devices who have not received a copy of the Safety
Alert or the Caregiver Information Sheet should contact their healthcare
provider, or Pulmonetic Systems to obtain a copy. Users with questions
should contact their healthcare providers. Healthcare providers with any
questions should contact the firm at 1-800-754-1914 extension
2.
Updated September 30, 2004