Jeff,

AS/NZS 3200.1.1 covers the safety requirements for medical electrical 
systems.

Below is a partial extract from standard which I hope helps to answer your 
email below. But please read the whole standard to obtain all the 
necessary information (click on 
http://www.saiglobal.com/shop/script/Details.asp?DocN=stds000013050 to 
obtain standard from Standards Australia) as the below as mentioned is a 
partial extract from the standard.

The note under the scope is very important because as it states the party 
who creates the system is responsible for making sure the system complies 
with the standard.

As a question to the BME community, is this being done in your department?

1 Scope and object
*1.201 Scope
This standard applies to the safety of MEDICAL ELECTRICAL SYSTEMS, as 
defined in 2.203. It
describes the safety requirements necessary to provide protection for the 
PATIENT, the
OPERATOR and surroundings.

NOTE ? It is presumed that the party assembling or modif ying the MEDICAL 
ELECTRICAL SYSTEMS will take the
necessary steps to assure compliance with this standard.


*2.203 MEDICAL ELECTRICAL SYSTEM (hereinafter referred to as a SYSTEM): 
Combinationof either
more than one item of MEDICAL ELECTRICAL EQUIPMENT or MEDICAL ELECTRICAL 
EQUIPMENT and
other non-medical electrical equipment having a specified function and 
inter-connected by:
- COUPLING, and/or
- aMULTIPLE PORTABLE SOCKET-OUTLET.
NOTE - The SYSTEM includes those accessories which are needed for 
operating the SYSTEM and are to be
specif ied by the manufacturer.
(See also examples given in annexes BBB and FFF.)

19 Continuous LEAKAGE CURRENTS and PATIENT AUXILIARY CURRENTS
*19.201 LEAKAGE CURRENTS
19.201.1 ENCLOSURE LEAKAGE CURRENT
In NORMAL CONDITION the allowable ENCLOSURE LEAKAGE CURRENT from or 
between parts of the
SYSTEM within the PATIENT ENVIRONMENT shall not exceed 0,1 mA.
In the event of the interruption of any non-permanently installed 
protective earth conductor the
allowable ENCLOSURE LEAKAGE CURRENT from or between parts of a SYSTEM 
within the PATIENT
ENVIRONMENT shall not exceed 0,5 mA. If the SYSTEM includes a MULTIPLE 
PORTABLE SOCKET- z
OUTLET the ENCLOSURE LEAKAGE CURRENT shall also be measured from parts 
which are in z
NORMAL CONDITION protectively earthed. z
19.201.2 PATIENT LEAKAGE CURRENT
In NORMAL CONDITION the PATIENT LEAKAGE CURRENT shall not exceed 0,1 mA 
for TYPE B and BF
EQUIPMENT and 0,01 mA for TYPE CF EQUIPMENT.
Compliance with the requirements of subclauses 19.201.1 and 19.201.2 is 
checked by
inspection and measurement of LEAKAGE CURRENTS using a device according to 
IEC 601-1,
subclause 19.4 e).
19.201.3 Connection of SIGNAL INPUT PARTS or SIGNAL OUTPUT PARTS (see also 
annex BBB)
If compliance of the MEDICAL ELECTRICAL EQUIPMENT with subclause 19.2 b) 
first dash
and/or 19.2 c) of IEC 601-1 is achieved by specifying that the SIGNAL 
INPUT PART and/or SIGNAL
OUTPUT PART is for exclusive connection to equipment as specified in the 
ACCOMPANYING
DOCUMENTS, then the SIGNAL INPUT PART and/or SIGNAL OUTPUT PART shall be 
connected:
- either to such equipment which, if CLASS I, is connected to the common 
PROTECTIVE
EARTH of the SYSTEM,
- ortoaSEPARATION DEVICE.
Compliance is checked by inspection.





Regards,
Scott Wedesweiler, Engineering Officer ? Electromedical

14 Nelson Street, Chatswood NSW 2067, Sydney, Australia
tel: 61 2 9410 5111 fax: 61 2 9415 1567
Testing & Certification Australia, a business of EnergyAustralia
Customer Comments/Feedback Form
P Please consider the environment before printing this email.

 





jeff.merri...@nt.gov.au 
21/05/2009 12:19 PM

To
"bmelist@bme.asn.au" <bmelist@bme.asn.au>
cc

Subject
[bmelist] RE: Earth Leakage







We have a central monitoring system comprising a PC, display and UPS in a
nurses station, central to the room and is also raised, which monitors 4 x
bedside patient monitors and the earth leakage for the central monitoring 
system
is 0.590 amps for normal and 1.161 amps for single fault (neutral open). 
The
note at the end of para 1.4.11 in AS/NZS 3551 states it is part of the 
medical
equipment system so therefore should be treated as such but it isn't in an 
area
that patients are treated . Any thaughts/remedies?

Regards
Jeff Merrigan
RDH Test & Commissioning Officer
Ph  08 89228423
Fax 08 89228712
email: "jeff.merri...@nt.gov.au"


To unsubscribe from the bmelist, send a blank email from the subscribed 
address to bmelist-unsubscr...@bme.asn.au




----------------------------------------------------------------------------------------
This e-mail may contain confidential or privileged information.   If you have 
received it in error, please notify the sender immediately via return e-mail 
and then delete the original e-mail. EnergyAustralia has collected your 
business contact details for dealing with you in your business capacity. More 
information about how we handle your personal information, including your right 
of access is contained at http://www.energy.com.au.

<<image/jpeg>>

Reply via email to