***  For details on how to be removed from this list visit the  ***
***          CCP4 home page http://www.ccp4.ac.uk         ***



Leif,

How do you measure sample temperature?

Do you have any references on the mobility of protons in hyperquenched water-cosolvent mixtures vs temperature? I have never heard 115K being quoted in this context before.

-James Holton
MAD Scientist

Hanson, Bryant wrote:

I feel like we are stealing thunder from Ana and Sean's session at ACA, but returning to the original question, going to a lower temperature (i.e. helium cooling) can increase the diffractive lifetime of your crystal. Experiments done by Unmesh Chinte under the direction of Connie Schall and I show that XI crystals will have their diffractive lifetime extended by a third if data measurement is done at 15K vs 100K. Paper is currently making the final rounds through the various authors prior to submission. It would get done quicker if we weren't all reading the BB.

Rant on temperature - never trust any cryostat to deliver at the crystal the temp it says on the readout. Eddie can say yea or nay on this but we have found that when the device says 100K the xtal is at a minimum 112K. Sometimes in the inverted orientation the temp will be 122K. This is important because protons will be mobile at 115K. So let's say 5 to 10K heating from the beam,at most beamlines you will be into the mobile proton region, and fast approaching the glass transition temperature. Again more work by Unmesh has shown there is a 1mm sweet spot in the center of a cold stream. Venturing outside will result in a 5 to 10K rise. 3mm away and local heating will cause the crystal to warm past the Tg.

So make sure you are centered in the cold stream during data measurement. As for using helium cooling, BioCARS and SBCCAT have hellium cryostats. The cost for a liter of liquid He in the US is less than a gallon of gas. Pinkerton's device uses about 4L per hour.

Leif Hanson

-----Original Message-----
From: [EMAIL PROTECTED] on behalf of Edward A. Berry
Sent: Tue 4/25/2006 4:13 PM
To: Ethan Merritt
Cc: James Holton; Ian Tickle; [EMAIL PROTECTED]; Edward Snell
Subject: Re: [ccp4bb]: Radiation damage problem

***  For details on how to be removed from this list visit the  ***
***          CCP4 home page http://www.ccp4.ac.uk         ***




Ethan Merritt wrote:
>
> Back of the lunch-time napkin thought experiment:
>
> 100 watt light bulb is exposed to air (essentially nitrogen), losing
> heat through radiation and conduction over a surface of 4piR^2 for R
> approximately 30 mm.  ===>  ~ .01 watt per square mm.
> The light bulb is hot, according to everyday experience

Say light bulb is 70*, or 50* above ambient
Heat dissipation is limitted by convection in the surrounding air,

>
> 100 micron cube protein crystal heated by 2% of a 30 milliwatt
> X-ray beam losing heat through conduction over a surface
> of 6 x 0.1^2mm  ===> ~ .01 watt per square mm.
> Crystal is now, one would think, light-bulb temperature.
>

Crystal might be 50* above ambient, but not light-bulb temperature.
And there is the wind-chill factor from the draft of the cold stream.

Still sounds like significant heating, but it is getting too
close for back-of the napkin calculations.

Ed