When we ran into reviewer complaints a while back we referred to Methods in Enzymology, 277:366-396. I can't lay my hands onto this to verify the statement, but back then it was good enough to placate the reviewer.

Full QnA with reviewer:

3) Why were only 1% of the reflections included in the test set for the free R value?

The original publication by Axel Brünger used 10% of the reflections for the test set, and this became standard practice. However, further analysis by Axel Brünger indicated that it is the absolute number of reflections in the test that is important, not the relative size, and he recommended that a test set size of at least 500 reflections (Methods in Enzymology, 277:366-396. Accordingly, we conservatively use a test set of 1,264 reflections, which gives the significant advantage that more terms can be used in the refinement calculation.


Andreas



Tim Gruene wrote:
On Tue, 17 Jun 2008, Anastassis Perrakis wrote:

been highly correlated with the phase accuracy of the atomic model. In
practice, about 5-10% of the observed diffraction data (chosen at random
from the unique reflections) become sequestered in the test set. The size
of the test set is a compromise between the desire to minimize statistical fluctuations of the free R value and the need to avoid a deleterious effect
on the atomic model by omission of too much experimental data.

I could not find the full text of the 1992 paper.

If I remember correctly (Tassos, you would have found it in the basement of the EMBL in Grenoble ;-)) Axel Brunger states in the 1992 article that 500-1000 reflections are sufficient for his purposes of Rfree calculations.

Tim