Thank you all for your informative responses!

While examining the effects of unusual beam profiles on data collection due to capillary optics, I had collected a wedge of data on a large, high-quality lysozyme crystal at 8 different sample to detector distances. I restricted the analysis of all data to the resolution range seen at the largest distance, so each data set contained nearly the same reflections with the same redundancy. <I/ sig> does indeed improve with distance: starting at 7.89 at 135 mm and reaching its maximum of 10.35 at 614 mm. The total linear R factor for that range also falls from 0.09 down to 0.059. So, even with a 1.8 mrad divergent beam (about as divergent as practical at a beamline), the effect is significant.

Surely someone has done this experiment before long ago.

I like Jim Plufgrath's way of looking at the problem: reflections and background scatter have different effective source sizes and distances. For a reflection, the source is the same as the beamline source (some place way upstream). For solvent the source is the sample itself, for air, the source is spread out over a range of distances ... thus background is very divergent.

Richard