There is always an anomalous signal to be seen if one measures things well
enough.  One does not need to be at the edge in order to detect anomalous
scatterers.  For example, most (if not all) of the sulfur SAD phasing is
done well away from the sulfur edge.
 
I will note that you stated that the f' and f" of Fe at this wavelength was
not zero, thus there should be a signal and a peak.  

When you phase your data, you will get better phases if you describe ALL the
anomalous scatterers and signals that are detectable and not just the Tb.
 
Jim

  _____  

From: CCP4 bulletin board [mailto:[email protected]] On Behalf Of
Huiming Li
Sent: Tuesday, August 09, 2011 11:38 AM
To: [email protected]
Subject: [ccp4bb] anomalous scatterer


Hi All,

  I am working on a Tb binding protein on which I collected anomalous data
at Tb edge of 1.648 A.  Each protein is designed to bind one Tb. There are
two copies of the protein in an ASU. I have two questions. First, I am only
able to see one copy of the protein with Tb bound, and no density on the
other copy.  Isn't this a bit surprising? Second, there is one additional
peak on each monomer at the site where Fe is known to bind, and Fe has an
edge of 1.739A. At 1.648A, f' and f'' of Fe is only about 1/5 of Tb. Is it
possible Fe also shows some anomalous signal at Tb edge?

Thank you,

Huiming Li, Ph.D.
Immune Disease Institute
Children's Hospital Boston
Harvard Medical School
Boston, MA 02115

Reply via email to