I'm glad George and Pavel have weighed in, since they're authorative figures.

I have another concern about the practice of using Fe2+ or Fe3+ scattering 
factors for iron in a heme.

One thing small molecule crystallographers do is make sure they/we get F000 
correct. This means that the number of electrons in the calculation must equal 
the number of electrons in the cell - easy for most small molecule structures 
but not for protein structures, and the easiest means of ensuring F000 is 
correct is to use neutral atom scattering factors (and, as George has pointed 
out, this does not significantly affect the final result).

Let's look at the iron heme complex as a whole. When uncoordinated to iron, the 
neutral macrocycle has two protons on the nitrogen atoms which dissociate when 
iron coordinates - the Fe2+ heme complex is neutral (ignoring whatever happens 
with the propionates).  So, if a crystallographer were  to use the ionic 
scattering factors for Fe, in order to get F000 correct, the crystallographer 
would need to add two electrons to the heme system.  It isn't clear where and 
how to do this in a small molecule structure let alone a protein structure 
(unless you're modeling the bonding density).

Sue

Sue Roberts
X-Ray Diffraction Facility, Department of Chemistry and Biochemistry
University of Arizona
[email protected]
________________________________
From: CCP4 bulletin board [[email protected]] on behalf of George Sheldrick 
[[email protected]]
Sent: Friday, June 20, 2014 12:41 AM
To: [email protected]
Subject: Re: [ccp4bb] refine an ion atom with different status

I agree with Pavel. Even for accurate small molecule data with R-values below 
3% the differences are hardly significant, the 'B-factors' compensate so well. 
The large majority of small molecule structures are refined with neutral atom 
scattering factors even if ions are present. The calculated scattering factor 
for an isolated ion in the gas phase is not really appropriate for the 
environement of an ion in the crystal anyway.

George


On 06/20/2014 07:18 AM, Pavel Afonine wrote:
Hi Bernhard,

phenix.refine makes use of charge if specified in PDB file (rightmost column 
after the chemical element type) to use appropriate form-factors. However, 
occupancies and B-factors are very efficient mops to accommodate a broad range 
of discrepancies between model and reality. So whether the effect of using 
charge is going to be noticeable, I guess, depends on the data quality 
(resolution, completeness, etc) and how strong the effect itself is.

Also, it should be relatively easy to make a numerical experiment with 
calculated data to see how the total scattering brakes down into individual 
contributions.

Pavel


On Thu, Jun 19, 2014 at 6:42 PM, Bernhard Rupp 
<[email protected]<mailto:[email protected]>> wrote:
“.. change the valence of ion or metal except by changing the occupancy”

Changing the occupancy is entirely different from changing valence. The former 
scales the scattering function proportionally, while the elimination of outer 
shell electrons predominantly reduces the very low resolution part (starting at 
f000) of the scattering function. Verifying the correct scattering function 
(e.g. Fe+++ vs Fe++ vs Fe atomic) used by the refinement program could be 
useful. I am curious: Garib, Pavel, Busters: How is that currently implemented?

Best, BR

From: CCP4 bulletin board 
[mailto:[email protected]<mailto:[email protected]>] On Behalf Of Wang, 
Bing
Sent: Thursday, June 19, 2014 10:44 PM
To: [email protected]<mailto:[email protected]>
Subject: [ccp4bb] refine an ion atom with different status

Hi CCP4 guys,

I have a structure with heme containing an ion atom in it. Except the 4 
coordinated nitrogen atoms in the heme, this ion also coordinates with one 
histidine residue and one ligand. But I found two negative red balls (top one 
and bottom one) around the ion, which is perpendicular to the heme plate and 
keeping in the same line with the histidine and my ligand (See the figure 
Ion_100 from coot in the attachment). I guess this ion has different status in 
it (e. g. mixture of Fe2+ and Fe3+). I simply tried the lower occupancy of ion. 
It clearly eliminate the negative ball at the bottom and most of the negative 
balls at the top, but also produced one more positive peak with slight movement 
instead of the negative ball at the bottom (See the figures Ion_90, Ion_85, 
Ion_80). The numbers in the image name represents the different occupancy 
("100" means 100%, "80" means 80%).

So any suggestions to solve this problem? Except changing the occupancy, is 
there a more precise way to change the valence of ion or metal in coot, and 
then refine in Refmac or Phenix?

Thanks!

Bing




--
Prof. George M. Sheldrick FRS
Dept. Structural Chemistry,
University of Goettingen,
Tammannstr. 4,
D37077 Goettingen, Germany
Tel. +49-551-39-33021 or -33068
Fax. +49-551-39-22582


Reply via email to