And the instrument makers (e.g., Rigaku and Bruker) currently sell multi-axis 
goniostats for in-house data collection--and not just for small-molecule 
purposes.

Diana

P.S. Ron, we still call 'em goniometer heads and goniostats at UT Southwestern.

> On Jul 14, 2017, at 1:00 PM, "[email protected]" 
> <[email protected]> wrote:
>
> I'm not understanding much of this discussion, but I've noticed the use of 
> words reminiscent of old issues concerning omega-two theta scans on 
> four-circle goniostats (see Stout and Jensen, pages 168-173, second edition).
>
> p.s.  In my upbringing, crystals were placed on goniometer heads so they 
> could be placed on goniometers or goniostats.
>
>> On Fri, 14 Jul 2017, Harry Powell wrote:
>>
>> hi folks
>> Just my two ha'porth - the small molecule crystallographers have been doing 
>> multi-orientation data collections
>> since they moved from point detectors to area detectors in the early 1990's, 
>> for the very reasons that Gerard
>> gives (their cusps are huge compared to ours...). Since they were perfectly 
>> used to using multi-axis goniostats*,
>> this wasn't a big psychological jump for them.
>> * I prefer calling the thing that the crystals sit on a "goniostat" because 
>> a "goniometer" is correctly something
>> for measuring angles (however, a "positioning goniometer" appears to be a 
>> specialised kind of goniostat);
>> wikipedia tells me that crystallographers seem to be the only group of 
>> people who confuse the two (but I didn't
>> read the article very carefully so IMWBW).
>> On 14 Jul 2017, at 15:15, Gerard Bricogne wrote:
>>      Dear Leo,
>>          What seems to have happened is that an existing thread where fine
>>      phi (actually: omega!) slicing was discussed, among many other things,
>>      digressed into a discussion of data collection protocols using more
>>      than one instrumental setting (either using a 2-theta motion of the
>>      detector, or a chi reorientation of the crystal). Briefly, my two
>>      cents on that topic: a 2-theta movement may help use different pixels
>>      on the detector, and could be valuable in filling the wide horizontal
>>      gaps on a Pilatus or Eiger, but it will leave the cusp in the same
>>      place and therefore will not fill it. Reorienting the crystal, on the
>>      other hand, can help cure all the known ills of single-sweep datasets
>>      (gaps and cusp in particular).
>>          On the matter of multi-orientation data collection, the idea and
>>      the practice go back (at least, in my memory) to Alan Wonacott, the
>>      co-creator of the Arndt-Wonacott rotation camera in the early 1970's.
>>      It was all done with gonio arcs. As each crystal had to be aligned
>>      manually in order to continue data collection where the previous one
>>      had left off, these arcs were in constant use, and there was always an
>>      extra cusp-filling collection at the end. Nowadays data collection has
>>      speeded up so much, and has become so dominated by automation, that
>>      multi-axis goniometry has been sidestepped because using it properly
>>      would have had to involve non-automated steps that are difficult to
>>      standardise (a notable exception being the protocol with 8 different
>>      values of Chi, using the PRIGo goniometer on the PX-III beamline at
>>      the SLS, that has been "instrumental" in enabling large structures to
>>      be experimentally phased by native SAD at 6keV).
>>          It is great to see that there are many developments underway in
>>      both hardware and software, leading gradually towards a reinstatement
>>      of multi-orientation data collection as an off-the-shelf option for
>>      those who are prepared to spend a bit more time to reliably get much
>>      better data. The Proxima-1 beamline scientists at SOLEIL have always
>>      been among the believers that the time would come when these efforts
>>      would bear fruit, and what my group has been able to do in this area
>>      owes a great deal to them.
>>          With best wishes,
>>               Gerard.
>>      --
>>      On Fri, Jul 14, 2017 at 01:18:35PM +0000, CHAVAS Leonard wrote:
>>            Reading back my email, when I mentioned 'just introduced', it is 
>> not giving justice to
>>            the reality and those who came up with the concept. I should have 
>> mentioned 'just
>>            reminded us', as the concept has been introduced quite a long 
>> time ago and few tens of
>>            communications. It is therefore a reminder that when coming to 
>> the will to collect good,
>>            clean and complete data, things aren't as simple as they would 
>> seem. Automation at our
>>            favourite beamlines do help by providing much more time thinking 
>> properly of the
>>            necessary strategies when coming to these difficult crystals so 
>> important to our hearts.
>>            Sorry again for the confusion. No hurt feelings I hope.
>>            Cheers, leo
>>            -
>>            Leonard Chavas
>>            -
>>            Synchrotron SOLEIL
>>            Proxima-I
>>            L'Orme des Merisiers
>>            Saint-Aubin - BP 48
>>            91192 Gif-sur-Yvette Cedex
>>            France
>>            -
>>            Phone:  +33 169 359 746
>>            Mobile: +33 644 321 614
>>            E-mail: [email protected]
>>            -
>>                  On 14 Jul 2017, at 14:07, CHAVAS Leonard
>>                  <[email protected]> wrote:
>>                  Just to comment on what Graeme just introduced. We (and I 
>> know we are not the
>>                  first ones and not the only ones) are pushing our user 
>> community towards this
>>                  procedure as a standard: lowering the transmission (less 
>> juicy, yet...) and
>>                  getting few data with various chi. It does help greatly in 
>> getting fully
>>                  complete data, with no loss in resolution. Just fantastic!
>>                  Cheers, leo
>>                  -
>>                  Leonard Chavas
>>                  -
>>                  Synchrotron SOLEIL
>>                  Proxima-I
>>                  L'Orme des Merisiers
>>                  Saint-Aubin - BP 48
>>                  91192 Gif-sur-Yvette Cedex
>>                  France
>>                  -
>>                  Phone:  +33 169 359 746
>>                  Mobile: +33 644 321 614
>>                  E-mail: [email protected]
>>                  -
>>                        On 14 Jul 2017, at 07:36, Graeme Winter
>>                        <[email protected]> wrote:
>>                        Jacob
>>                        If you have a complete 360 deg data set and your 
>> sample is still
>>                        alive, and you have a multi-axis gonio, I would 
>> recommend
>>                        rotating the crystal about the beam (ideally by ~ 
>> maximum
>>                        scattering 2-theta angle) and collecting again. This 
>> would record
>>                        your blind region as well as moving the reflections 
>> to different
>>                        pixels, and (as a bonus) also will move reflections 
>> out from the
>>                        tile join regions into somewhere they can be 
>> measured, which
>>                        would not happen for small 2-theta shift.
>>                        See http://scripts.iucr.org/cgi-bin/paper?BA0020 
>> Figure 16 as
>>                        excellent illustration of this.
>>                        Biggest risk with this is getting *moving* shadows on 
>> the data on
>>                        the second run, as an effective 45-50 degree chi 
>> shift (say) will
>>                        usually be a pretty wide opening angle for a kappa 
>> gonio. XDS and
>>                        DIALS both have mechanisms to deal with this, and 
>> automated
>>                        processing packages are able to apply these given a 
>> reasonable
>>                        understanding of the beamline.
>>                        Also saves building 2-theta axes which can handle 92 
>> kg ;o)
>>                        Cheers Graeme
>>                        On 13 Jul 2017, at 21:00, Keller, Jacob
>>                        
>> <[email protected]<mailto:[email protected]>>
>>                        wrote:
>>                        I thought there was a new paper from the Pilatus 
>> people saying
>>                        fine slicing is worth it even beyond the original 1/2 
>> mosaicity
>>                        rule?
>>                        I would think, actually, more gains would made by 
>> doing light
>>                        exposures at, say, 1/3 mosaicity, collecting 360 deg, 
>> then
>>                        shifting the detector in 2theta by a degree or two to 
>> shift
>>                        uniformly the spots to new pixels, maybe accompanied 
>> by a kappa
>>                        change. One would have to remember about the 
>> two-theta when
>>                        processing, however!
>>                        JPK
>>                        -----Original Message-----
>>                        From: CCP4 bulletin board 
>> [mailto:[email protected]] On
>>                        Behalf Of Gerd Rosenbaum
>>                        Sent: Thursday, July 13, 2017 3:40 PM
>>                        To: 
>> [email protected]<mailto:[email protected]>
>>                        Subject: Re: [ccp4bb] weird diffraction pattern
>>                        Dear Gerard,
>>                        my "sound like a sales person" was meant as poking a 
>> little fun -
>>                        nothing serious, of course.
>>                        I and our users like our not-so-new-anymore Pilatus3 
>> 6M. It's a
>>                        great detector in many ways. But, there is a lot of 
>> hype that
>>                        this detector solves all-problem, for instance fine 
>> slicing that
>>                        is claimed to be only possible with a pixel array 
>> detector.
>>                        People get carried away and use
>>                        0.01 degree slices even as the mosaicity of their 
>> sample is, say,
>>                        0.3 degree. Slicing beyond 1/3 of the mosaicity will 
>> gain you
>>                        very little - only more frames, more processing time.
>>                        This discourse is already drifting away from the 
>> original topic
>>                        of the thread so I will comment on the other 
>> arguments  you made
>>                        like resolution in a private e-mail.
>>                        Best regards,
>>                        Gerd
>>                        On 13.07.2017 14:00, Gerard Bricogne wrote:
>>                        Dear Gerd,
>>                          I can assure you that I have no shares in Dectris 
>> nor any
>>                        commecial connections with them. What I do have is a 
>> lot of still
>>                        vivid memories of CCD images, with their wooly 
>> point-spread
>>                        function
>>                        that was affected by fine-grained spatial variability 
>> as well as
>>                        by
>>                        irredicible inaccuracies in the geometric corrections 
>> required to
>>                        try
>>                        and undo the distortions introduced by the 
>> fiber-optic taper. By
>>                        comparison the pixel-array detectors have a very 
>> regular
>>                        structure, so
>>                        that slight deviations from exact registering of the 
>> modules can
>>                        be
>>                        calibrated with high accuracy, making it possible to 
>> get very
>>                        small
>>                        residuals between calculated and observed spot 
>> positions. That, I
>>                        certainly never saw with CCD images.
>>                          I do think that asking for the image width was a 
>> highly
>>                        pertinent question in this case, that had not been 
>> asked. As a
>>                        specialist you might know how to use a CCD to good 
>> effect in
>>                        fine-slicing mode, but it is amazing how many people 
>> there are
>>                        still
>>                        out there who are told to use 0.5 or even 1.0 degree 
>> image
>>                        widths.
>>                          Compensating the poor PSF of a CCD by fine slicing 
>> in the
>>                        angular dimension is a tall order. With a Pilatus at 
>> 350mm from
>>                        the
>>                        crystal, the angular separation between 174-micron 
>> pixels is 0.5
>>                        milliradian.
>>                        To achieve that separation in the angular (rotation) 
>> dimension,
>>                        the
>>                        equivalent image width would have to be 0.03 degree. 
>> For an EIGER
>>                        the
>>                        numbers become 75 microns, hence 0.21 milliradian 
>> i.e. 0.012
>>                        degree.
>>                          Hence my advice, untainted by any commercial agenda 
>> :-) .
>>                          With best wishes,
>>                               Gerard.
>>                        --
>>                        On Thu, Jul 13, 2017 at 01:25:08PM -0500, Gerd 
>> Rosenbaum wrote:
>>                        Dear Gerard,
>>                        you sound like a sales person for Dectris. Fine 
>> slicing is
>>                        perfectly
>>                        fine with CCD detectors - it takes a bit longer 
>> because of the
>>                        step
>>                        scan instead of continuous scan. The read noise issue 
>> is often
>>                        overstated compared to the sample induced scatter 
>> background. If
>>                        for
>>                        fine slicing at 0.05 degree or less the diffraction 
>> peaks go too
>>                        close to the read noise make a longer exposure - 
>> signal goes up,
>>                        ratio signal to sample-induced-BG less, as for any 
>> fine slicing,
>>                        same read noise.
>>                        It would be helpful to analyze the dense spot packing 
>> along layer
>>                        lines if we knew the wavelength and the 
>> sample-to-detector
>>                        distance
>>                        (assuming this is a 300 mm detector) and the rotation 
>> width - as
>>                        you
>>                        pointed out. That would help to distinguish between 
>> multiple
>>                        crystals
>>                        (my guess) and lattice translocation disorder. Fine 
>> slicing is
>>                        definitely needed to figure out what the diffraction 
>> pattern at
>>                        120
>>                        degree could tell you in terms of strong anisotropy .
>>                        Best regard.
>>                        Gerd
>>                        On 13.07.2017 08:20, Gerard Bricogne wrote:
>>                        Dear Tang,
>>                          I noticed that your diffraction images seem to have 
>> been
>>                        recorded on a 3x3 CCD detector. With this type of 
>> detector, fine
>>                        slicing is often discouraged (because of the readout 
>> noise), and
>>                        yet
>>                        with the two long cell axes you have, any form of 
>> thick (or only
>>                        semi-fine) slicing would result in spot overlaps.
>>                          What, then, was your image width? Would you have 
>> access to a
>>                        beamline with a Pilatus detector so that you could 
>> collect
>>                        fine-sliced data?
>>                          I would tend to agree with Herman that your 
>> crystals might be
>>                        cursed with lattice translocation disorder (LTD), but 
>> you might
>>                        as
>>                        well try and put every chance of surviving this on 
>> your side by
>>                        making sure that you collect fine-sliced data. LTD 
>> plus thick
>>                        slicing would give you random data along the streaky 
>> direction.
>>                        Use
>>                        an image width of at most 0.1 degree (0.05 would be 
>> better) on a
>>                        Pilatus, and use XDS to process your images.
>>                          Good luck!
>>                            Gerard
>>                        --
>>                        On Thu, Jul 13, 2017 at 01:21:02PM +0100, Tang 
>> Chenjun wrote:
>>                        Hi David,
>>                        Thanks for your comments. Although the spots become 
>> streaky in
>>                        certain directions, I have processed the data in 
>> HKL3000 and
>>                        imosflm, which suggested the C2221 space group 
>> (66.59, 246.95 and
>>                        210.17). The Rmerge(0.14), completeness(94.8%), 
>> redundancy(4.6)
>>                        are OK. When I tried to run Balbes with the solved 
>> native
>>                        structure, the molecular replacement solution was 
>> poor. So I ran
>>                        Balbes with the split domains of the native 
>> structure. Although
>>                        the solutions were also poor, I found the MR score of 
>> one
>>                        solution above 35. On the basis of this solution, I 
>> tried to run
>>                        Buccaneer and the Rfree could be 0.46. Unfortunately, 
>> there are
>>                        four molecules in the asymmetric unit and it is to 
>> hard for me to
>>                        reduce the Rfree further.
>>                        All best,
>>                        Chenjun Tang
>>                        --
>>                        This e-mail and any attachments may contain 
>> confidential,
>>                        copyright and or privileged material, and are for the 
>> use of the
>>                        intended addressee only. If you are not the intended 
>> addressee or
>>                        an authorised recipient of the addressee please 
>> notify us of
>>                        receipt by returning the e-mail and do not use, copy, 
>> retain,
>>                        distribute or disclose the information in or attached 
>> to the
>>                        e-mail.
>>                        Any opinions expressed within this e-mail are those 
>> of the
>>                        individual and not necessarily of Diamond Light 
>> Source Ltd.
>>                        Diamond Light Source Ltd. cannot guarantee that this 
>> e-mail or
>>                        any attachments are free from viruses and we cannot 
>> accept
>>                        liability for any damage which you may sustain as a 
>> result of
>>                        software viruses which may be transmitted in or with 
>> the message.
>>                        Diamond Light Source Limited (company no. 4375679). 
>> Registered in
>>                        England and Wales with its registered office at 
>> Diamond House,
>>                        Harwell Science and Innovation Campus, Didcot, 
>> Oxfordshire, OX11
>>                        0DE, United Kingdom
>> Harry
>> --
>> Dr Harry Powell