In my opinion the threshold should be zero bits.  Yes, this is where CC1/2 = 0 
(or FSC = 0).  If there is correlation then there is information, and why throw 
out information if there is information to be had?  Yes, this information comes 
with noise attached, but that is why we have weights.

It is also important to remember that zero intensity is still useful 
information.  Systematic absences are an excellent example.  They have no 
intensity at all, but they speak volumes about the structure.  In a similar 
way, high-angle zero-intensity observations also tell us something.  Ever tried 
unrestrained B factor refinement at poor resolution?  It is hard to do nowadays 
because of all the safety catches in modern software, but you can get great R 
factors this way.  A telltale sign of this kind of "over fitting" is remarkably 
large Fcalc values beyond the resolution cutoff.  These don't contribute to the 
R factor, however, because Fobs is missing for these hkls. So, including 
zero-intensity data suppresses at least some types of over-fitting.

The thing I like most about the zero-information resolution cutoff is that it 
forces us to address the real problem: what do you mean by "resolution" ?  Not 
long ago, claiming your resolution was 3.0 A meant that after discarding all 
spots with individual I/sigI < 3 you still have 80% completeness in the 3.0 A 
bin.  Now we are saying we have a 3.0 A data set when we can prove 
statistically that a few non-background counts fell into the sum of all spot 
areas at 3.0 A.  These are not the same thing.

Don't get me wrong, including the weak high-resolution information makes the 
model better, and indeed I am even advocating including all the noisy zeroes.  
However, weak data at 3.0 A is never going to be as good as having strong data 
at 3.0 A.  So, how do we decide?  I personally think that the resolution 
assigned to the PDB deposition should remain the classical I/sigI > 3 at 80% 
rule.  This is really the only way to have meaningful comparison of resolution 
between very old and very new structures.  One should, of course, deposit all 
the data, but don't claim that cut-off as your "resolution".  That is just 
plain unfair to those who came before.

Oh yeah, and I also have a session on "interpreting low-resolution maps" at the 
GRC this year.  
https://www.grc.org/diffraction-methods-in-structural-biology-conference/2020/<https://urldefense.com/v3/__https://www.grc.org/diffraction-methods-in-structural-biology-conference/2020/__;!!Eh6p8Q!TK-tIY-zm5coRu74uWMkIJkTFWNz4-1ibr1oaahxT_2BAAetUTMNdfRqUCmIqtYfhDo$>

So, please, let the discussion continue!

-James Holton
MAD Scientist

On 2/22/2020 11:06 AM, Nave, Colin (DLSLtd,RAL,LSCI) wrote:
Alexis
This is a very useful summary.

You say you were not convinced by Marin's derivation in 2005. Are you convinced 
now and, if not, why?

My interest in this is that the FSC with half bit thresholds have the danger of 
being adopted elsewhere because they are becoming standard for protein 
structure determination (by EM or MX). If it is used for these mature 
techniques it must be right!

It is the adoption of the ½ bit threshold I worry about. I gave a rather weak 
example for MX which consisted of partial occupancy of side chains, substrates 
etc. For x-ray imaging a wide range of contrasts can occur and, if you want to 
see features with only a small contrast above the surroundings then I think the 
half bit threshold would be inappropriate.

It would be good to see a clear message from the MX and EM communities as to 
why an information content threshold of ½ a bit is generally appropriate for 
these techniques and an acknowledgement that this threshold is 
technique/problem dependent.

We might then progress from the bronze age to the iron age.

Regards
Colin



From: CCP4 bulletin board <CCP4BB@JISCMAIL.AC.UK><mailto:CCP4BB@JISCMAIL.AC.UK> 
On Behalf Of Alexis Rohou
Sent: 21 February 2020 16:35
To: CCP4BB@JISCMAIL.AC.UK<mailto:CCP4BB@JISCMAIL.AC.UK>
Subject: Re: [ccp4bb] [3dem] Which resolution?

Hi all,

For those bewildered by Marin's insistence that everyone's been messing up 
their stats since the bronze age, I'd like to offer what my understanding of 
the situation. More details in this thread from a few years ago on the exact 
same topic:
https://mail.ncmir.ucsd.edu/pipermail/3dem/2015-August/003939.html<https://urldefense.com/v3/__https://mail.ncmir.ucsd.edu/pipermail/3dem/2015-August/003939.html__;!!Eh6p8Q!TK-tIY-zm5coRu74uWMkIJkTFWNz4-1ibr1oaahxT_2BAAetUTMNdfRqUCmIsJF61uc$>
https://mail.ncmir.ucsd.edu/pipermail/3dem/2015-August/003944.html<https://urldefense.com/v3/__https://mail.ncmir.ucsd.edu/pipermail/3dem/2015-August/003944.html__;!!Eh6p8Q!TK-tIY-zm5coRu74uWMkIJkTFWNz4-1ibr1oaahxT_2BAAetUTMNdfRqUCmIPu-nRBo$>

Notwithstanding notational problems (e.g. strict equations as opposed to 
approximation symbols, or omission of symbols to denote estimation), I believe 
Frank & Al-Ali and "descendent" papers (e.g. appendix of Rosenthal & Henderson 
2003) are fine. The cross terms that Marin is agitated about indeed do in fact 
have an expectation value of 0.0 (in the ensemble; if the experiment were 
performed an infinite number of times with different realizations of noise). I 
don't believe Pawel or Jose Maria or any of the other authors really believe 
that the cross-terms are orthogonal.

When N (the number of independent Fouier voxels in a shell) is large enough, 
mean(Signal x Noise) ~ 0.0 is only an approximation, but a pretty good one, 
even for a single FSC experiment. This is why, in my book, derivations that 
depend on Frank & Al-Ali are OK, under the strict assumption that N is large. 
Numerically, this becomes apparent when Marin's half-bit criterion is plotted - 
asymptotically it has the same behavior as a constant threshold.

So, is Marin wrong to worry about this? No, I don't think so. There are indeed 
cases where the assumption of large N is broken. And under those circumstances, 
any fixed threshold (0.143, 0.5, whatever) is dangerous. This is illustrated in 
figures of van Heel & Schatz (2005). Small boxes, high-symmetry, small objects 
in large boxes, and a number of other conditions can make fixed thresholds 
dangerous.

It would indeed be better to use a non-fixed threshold. So why am I not using 
the 1/2-bit criterion in my own work? While numerically it behaves well at most 
resolution ranges, I was not convinced by Marin's derivation in 2005. 
Philosophically though, I think he's right - we should aim for FSC thresholds 
that are more robust to the kinds of edge cases mentioned above. It would be 
the right thing to do.

Hope this helps,
Alexis



On Sun, Feb 16, 2020 at 9:00 AM Penczek, Pawel A 
<pawel.a.penc...@uth.tmc.edu<mailto:pawel.a.penc...@uth.tmc.edu>> wrote:
Marin,

The statistics in 2010 review is fine. You may disagree with assumptions, but I 
can assure you the “statistics” (as you call it) is fine. Careful reading of 
the paper would reveal to you this much.
Regards,
Pawel


On Feb 16, 2020, at 10:38 AM, Marin van Heel 
<marin.vanh...@googlemail.com<mailto:marin.vanh...@googlemail.com>> wrote:


**** EXTERNAL EMAIL ****
Dear Pawel and All others ....
This 2010 review is - unfortunately - largely based on the flawed statistics I 
mentioned before, namely on the a priori assumption that the inner product of a 
signal vector and a noise vector are ZERO (an orthogonality assumption).  The 
(Frank & Al-Ali 1975) paper we have refuted on a number of occasions (for 
example in 2005, and most recently in our BioRxiv paper) but you still take 
that as the correct relation between SNR and FRC (and you never cite the 
criticism...).
Sorry
Marin

On Thu, Feb 13, 2020 at 10:42 AM Penczek, Pawel A 
<pawel.a.penc...@uth.tmc.edu<mailto:pawel.a.penc...@uth.tmc.edu>> wrote:
Dear Teige,

I am wondering whether you are familiar with

Resolution measures in molecular electron microscopy.
Penczek PA. Methods Enzymol. 2010.
Citation

Methods Enzymol. 2010;482:73-100. doi: 10.1016/S0076-6879(10)82003-8.

You will find there answers to all questions you asked and much more.

Regards,
Pawel Penczek

Regards,
Pawel
_______________________________________________
3dem mailing list
3...@ncmir.ucsd.edu<mailto:3...@ncmir.ucsd.edu>
https://mail.ncmir.ucsd.edu/mailman/listinfo/3dem<https://urldefense.proofpoint.com/v2/url?u=https-3A__mail.ncmir.ucsd.edu_mailman_listinfo_3dem&d=DwMFaQ&c=bKRySV-ouEg_AT-w2QWsTdd9X__KYh9Eq2fdmQDVZgw&r=yEYHb4SF2vvMq3W-iluu41LlHcFadz4Ekzr3_bT4-qI&m=3-TZcohYbZGHCQ7azF9_fgEJmssbBksaI7ESb0VIk1Y&s=XHMq9Q6Zwa69NL8kzFbmaLmZA9M33U01tBE6iAtQ140&e=>
_______________________________________________
3dem mailing list
3...@ncmir.ucsd.edu<mailto:3...@ncmir.ucsd.edu>
https://mail.ncmir.ucsd.edu/mailman/listinfo/3dem<https://urldefense.com/v3/__https://mail.ncmir.ucsd.edu/mailman/listinfo/3dem__;!!Eh6p8Q!TK-tIY-zm5coRu74uWMkIJkTFWNz4-1ibr1oaahxT_2BAAetUTMNdfRqUCmI7LD77u4$>

________________________________

To unsubscribe from the CCP4BB list, click the following link:
https://www.jiscmail.ac.uk/cgi-bin/webadmin?SUBED1=CCP4BB&A=1<https://urldefense.com/v3/__https://www.jiscmail.ac.uk/cgi-bin/webadmin?SUBED1=CCP4BB&A=1__;!!Eh6p8Q!TK-tIY-zm5coRu74uWMkIJkTFWNz4-1ibr1oaahxT_2BAAetUTMNdfRqUCmI1pndYoE$>



--

This e-mail and any attachments may contain confidential, copyright and or 
privileged material, and are for the use of the intended addressee only. If you 
are not the intended addressee or an authorised recipient of the addressee 
please notify us of receipt by returning the e-mail and do not use, copy, 
retain, distribute or disclose the information in or attached to the e-mail.
Any opinions expressed within this e-mail are those of the individual and not 
necessarily of Diamond Light Source Ltd.
Diamond Light Source Ltd. cannot guarantee that this e-mail or any attachments 
are free from viruses and we cannot accept liability for any damage which you 
may sustain as a result of software viruses which may be transmitted in or with 
the message.
Diamond Light Source Limited (company no. 4375679). Registered in England and 
Wales with its registered office at Diamond House, Harwell Science and 
Innovation Campus, Didcot, Oxfordshire, OX11 0DE, United Kingdom


________________________________

To unsubscribe from the CCP4BB list, click the following link:
https://www.jiscmail.ac.uk/cgi-bin/webadmin?SUBED1=CCP4BB&A=1<https://urldefense.com/v3/__https://www.jiscmail.ac.uk/cgi-bin/webadmin?SUBED1=CCP4BB&A=1__;!!Eh6p8Q!TK-tIY-zm5coRu74uWMkIJkTFWNz4-1ibr1oaahxT_2BAAetUTMNdfRqUCmI1pndYoE$>


________________________________

To unsubscribe from the CCP4BB list, click the following link:
https://www.jiscmail.ac.uk/cgi-bin/webadmin?SUBED1=CCP4BB&A=1<https://urldefense.com/v3/__https://www.jiscmail.ac.uk/cgi-bin/webadmin?SUBED1=CCP4BB&A=1__;!!Eh6p8Q!TK-tIY-zm5coRu74uWMkIJkTFWNz4-1ibr1oaahxT_2BAAetUTMNdfRqUCmI1pndYoE$>

########################################################################

To unsubscribe from the CCP4BB list, click the following link:
https://www.jiscmail.ac.uk/cgi-bin/webadmin?SUBED1=CCP4BB&A=1

Reply via email to