Dear Sergei,

Sorry to hear that you have lost samples due to the dry shipper warming up.

In all our shipments to/from various synchrotrons we've only had two arrive 
'warm', shipping to/from London.

One was returning from the Swiss Light Source to London and where the vacuum 
failed during transit, so it arrived warm. We disposed of this dry shipper 
rather than trying to re-evacuate it.

The second was en-route to Diamond Light Source where the courier had the box 
over the weekend at their intermediate storage at Heathrow. We assumed that it 
was laid on its side from Friday - Monday as the vacuum was still fine and the 
shipper retained nitrogen and is still in use, but the samples for this 
shipment were lost.

These failed shipments have been over a period of many years with many 
shipments per year (>50 if counting out and back separately) to various sites 
around Europe and a handful to the USA.

We've had another shipper where the vacuum failed overnight whilst it was still 
in the lab and full of liquid nitrogen. Fortunately, I managed to rescue the 
samples into a second shipper with no apparent issue. The sign here was the 
outside being covered with condensation and being cold to the touch.

We generally use either the square plastic shipping boxes or the smaller 
metallic ones, but we do have one dome shaped one. Nearly all our shippers are 
CXR100, as it is easier to load Unipucks, and we try and test the retention 
time at least one or twice a year.

Couriers have been a combination of DHL to Diamond in the UK and Fedex to/from 
Europe. In the past, we've only used more expensive specialist couriers for 
same day shipments.

It might be worth checking that the local courier is keeping the shipper 
upright for any overnight storage. Following the issue above, we generally ship 
early in the week to minimise time where the dewars being stored over the 
weekend.

We have had deliveries of equipment with tilt indicators, and they might act as 
a deterrent as well.

Hope this is of help.

Andy Purkiss
Crystallography Lab Manager.
________________________________
From: CCP4 bulletin board <[email protected]> on behalf of Sergei Strelkov 
<[email protected]>
Sent: 22 October 2026 14:42
To: [email protected] <[email protected]>
Subject: Re: [ccp4bb] About dry dewars

You don't often get email from [email protected]. 
Learn why this is important<https://aka.ms/LearnAboutSenderIdentification>

External Sender: Use caution.

Dear Community,

I wanted to pick up this thread as we recently had two instances of our 
crystals arriving warm to the synchrotron.

Our shipments are within continental Europe to the ESRF and take under two days 
if we are lucky, but often longer. We ship dewars in standard shipping cases.

We did the following tests with our not-that-new CX100 and also CXR100 (a 
wider-neck version of CX100).
The dewars were saturated with LN2, with cold racks/pucks in place.
When stored in the normal upright position, both our dewars stayed cold for 10 
days or more.
However, when we laid them on the side, the hold time was only about two days 
for CX100 and less than 24h in some tests with CXR100!

Our conclusion is that the problem lays in the courier not storing the dewars 
properly in the upright position.

As we are getting a bit desperate, I would appreciate your opinions/advice.
Should we get a special dome-shaped enclosure like Upright Shipping case for 
CX100 and CXR100, Dry Shippers - Jena 
Bioscience<https://www.jenabioscience.com/crystallography-cryo-em/specimen-storage-shipping/dry-shippers/x-4lsc-upright-shipping-case-for-cx100-and-cxr100>
 ?
Should we attach a tilt indicator to our existing enclosure? (In addition to 
the standard 'keep upright' sign which is there already of course)
Should we use a dedicated expensive courier for scientific samples?

Thank you!
Sergei




Prof. Sergei V. Strelkov

Laboratory for Biocrystallography, Department of Pharmaceutical and 
Pharmacological Sciences, KU Leuven

O&N2, Campus Gasthuisberg, Herestraat 49 bus 822, 3000 Leuven, Belgium

Phone: +32 486 29 41 32 (mobile)

Lab pages: https://gbiomed.kuleuven.be/english/research/50000715/50000719/

________________________________
From: CCP4 bulletin board <[email protected]> on behalf of Mitchell D. 
Miller <[email protected]>
Sent: Friday, October 18, 2024 16:48
To: [email protected] <[email protected]>
Subject: Re: [ccp4bb] About dry dewars

Another good way to test the dry shipping dewar is to its dry/empty weight.
If you don't know your "as new" mass, you can weigh it, then
purge with dry air/nitrogen for 24 hours and check the weight again,
if it is unchanged, then you can use that as your dry/empty weight.

Next you can charge the shipping dewar according to the manufacturer's
specs. Pour out the excess LN2. Then weigh the unit. This will tell
you the capacity of LN2 retained in the absorbent material. (You can
compare this to the manufactures specs). Then you can leave the unit
for 24 hours and then weigh it again.  From the difference of the mass,
you can calculate the nominal evaporation rate. E.g. how much LN2
is lost each day.  Then divide the mass of the retained LN2
by this evaporation rate to find the number of days of holding
time.

For a CX100 dewar, the absorbent material should hold  4.4 L of LN2
which is 3.6 kg and should have a holding time of ~21 days when new.
For the MVE SC4/2 it holds 3.8 L (3.1 kg) with an expected holding
time of 14 days when new. (note 1 l LN2 = 0.808 kg)

A good check you can do each time before sending out your samples
is to fill the shipper the day before (you can leave excess liquid
in the shipper). After you have fully charged it and it is cold (an
hour or so) record the weight. Then take another measurement 12-24 hrs
later and calculate your evaporation rate to confirm it is close to
what you expect.

There are some nice guides at
https://smb.slac.stanford.edu/facilities/hardware/cryotools/shipping-dewar-testing.html
https://www.diamond.ac.uk/Instruments/Mx/Common/Common-Manual/Shipping-Samples/Care-of-dry-shippers.html

I know that the LS-CAT group at the APS has the correct vacuum operator
to pump out the CX100 dewars. If that is the same one you have, they
might be able to tell you the source/supplier for it. I had looked for
the operator for the MVE SC4/2 type dewars, but did not find confirmation
on the correct thread pitch. When I contacted tech support at the
vendor we purchased the dewar from, they contacted the manufacture
and were told that they don't consider it user serviceable and would not
relay the diameter and thread pitches.

Here is a general video on the evacuation process --
https://www.youtube.com/watch?v=hnPZrGdcTSs
Here is a DIY solution ---
http://www.nuclearphysicslab.com/npl/npl-home/spectroscopy/software_and_hardware/refurbishment-of-ln2-dewars/


Regards,
Mitch


Quoting Artem Evdokimov <[email protected]>:

> Cheapest possible system to check dry shipper would be a wooden stick
> lowered into the shipper then taken out and exposed to moist air.
> Condensation would indicate a cold zone...
>
> Second cheapest option is a red LED on a pair of long thin wires. If it
> changes to green as it dips into the dewar, again- it is cold in there. The
> change will be partial depending on the actual temperature. Please note
> this only works with inorganic LEDs.
>
> If you crave precision and actual numbers, then you should invest a dollar
> or two in a type K thermocouple. It is good down to -200 C which is lower
> than LN2 temperature. These are on ebay and Amazon for a few bucks. A few
> more bucks will get you a thermometer body with the right connector, or you
> can use a multimeter and recall that the voltage change should be about 41
> microcolts per degree C. You can calibrate your system in LN2 and ice or
> boiling water to get the correct thermal coefficient and offset. Just don't
> forget to use the same connection and instrument all the time since the
> reference junction will change if you change the setup.
>
> As to re-evacuating the dewar - this varies depending on how your
> particular device is constructed. Some have a 'nipple' style valve, others
> have a needle style valve and some are permanently welded shut... it's not
> obvious that your mechanics shop will have the requisite vacuum connector
> to service your shipper. We used to send ours to manufacturer for
> reconditioning.
>
> Best of luck,
>
> Artem
>
> On Fri, Oct 18, 2024, 20:23 Pedro Matias <[email protected]> wrote:
>
>> Dear All,
>>
>> I'm looking for a low temperature measuring system to check our dry
>> dewars, preferably one that is not very expensive. Any recommendations ?
>>
>> Also, I was told that it is possible to regenerate the vacuum in those
>> vessels. Does anybody have any experience with this? If so, how can it
>> be done?
>>
>> Thanks in advance,
>>
>> Pedro Matias
>>
>> --
>> Industry and Medicine Applied Crystallography
>> Macromolecular Crystallography Unit
>> ___________________________________
>> Phones : (351-21) 446-9100 Ext. 1669
>>           (351-21) 446-9669 (direct)
>>   Fax   : (351-21) 441-1277 or 443-3644
>>
>> email : [email protected]
>>
>>
>> http://www.itqb.unl.pt/research/biological-chemistry/industry-and-medicine-applied-crystallography
>> http://www.itqb.unl.pt/labs/macromolecular-crystallography-unit
>>
>> Mailing address :
>> Instituto de Tecnologia Quimica e Biologica António Xavier
>> Universidade Nova de Lisboa
>> Av. da República
>> 2780-157 Oeiras
>> PORTUGAL
>>
>> ITQB NOVA, a great choice for your PhD
>> https://youtu.be/de6j-aaTWNQ
>>
>> Master Programme in Biochemistry for Health
>> https://youtu.be/UKstDCFjYI8
>>
>> ########################################################################
>>
>> To unsubscribe from the CCP4BB list, click the following link:
>> https://www.jiscmail.ac.uk/cgi-bin/WA-JISC.exe?SUBED1=CCP4BB&A=1
>>
>> This message was issued to members of 
>> www.jiscmail.ac.uk/CCP4BB<http://www.jiscmail.ac.uk/CCP4BB>, a
>> mailing list hosted by www.jiscmail.ac.uk<http://www.jiscmail.ac.uk>, terms 
>> & conditions are
>> available at https://www.jiscmail.ac.uk/policyandsecurity/
>>
>
> ########################################################################
>
> To unsubscribe from the CCP4BB list, click the following link:
> https://www.jiscmail.ac.uk/cgi-bin/WA-JISC.exe?SUBED1=CCP4BB&A=1
>
> This message was issued to members of 
> www.jiscmail.ac.uk/CCP4BB<http://www.jiscmail.ac.uk/CCP4BB>, a
> mailing list hosted by www.jiscmail.ac.uk<http://www.jiscmail.ac.uk>, terms & 
> conditions are
> available at https://www.jiscmail.ac.uk/policyandsecurity/

########################################################################

To unsubscribe from the CCP4BB list, click the following link:
https://www.jiscmail.ac.uk/cgi-bin/WA-JISC.exe?SUBED1=CCP4BB&A=1

This message was issued to members of 
www.jiscmail.ac.uk/CCP4BB<http://www.jiscmail.ac.uk/CCP4BB>, a mailing list 
hosted by www.jiscmail.ac.uk<http://www.jiscmail.ac.uk>, terms & conditions are 
available at https://www.jiscmail.ac.uk/policyandsecurity/

________________________________

To unsubscribe from the CCP4BB list, click the following link:
https://www.jiscmail.ac.uk/cgi-bin/WA-JISC.exe?SUBED1=CCP4BB&A=1

The Francis Crick Institute Limited is a registered charity in England and 
Wales no. 1140062 and a company registered in England and Wales no. 06885462, 
with its registered office at 1 Midland Road London NW1 1AT

########################################################################

To unsubscribe from the CCP4BB list, click the following link:
https://www.jiscmail.ac.uk/cgi-bin/WA-JISC.exe?SUBED1=CCP4BB&A=1

This message was issued to members of www.jiscmail.ac.uk/CCP4BB, a mailing list 
hosted by www.jiscmail.ac.uk, terms & conditions are available at 
https://www.jiscmail.ac.uk/policyandsecurity/

Reply via email to