Thanks for doing the TX side one tone and two tone tests for USRPN210. It
helps me get a reference.

1) When I did the test for finding out the calibration factor, I found it
was a constant value ntil a certain range and after that it dipped (e.g ard
-8 dBm input power @ 400 Mhz). I image this is point near the saturation
/compression region of the device. This was observed on 5 USRPN210+WBX
devices that I tested.
a) So I understand that this calibration factor is valid only in the linear
range. Is it correct?
b) if so, I should not use the calibration factor to turn the IMD products'
value on FFT to dBm using this factor as they occur only when the device is
operating in the non linear range.

Please let me know if my understanding is right.

2) You told me to plot the IMD plot with difference of power between
fundamental tone and the IMD product power @ harmonc freq. i.e  [power @ F1
- power @ (2F1-F2)] . But this is coming as a +ve value for me while your
plot shows the Y axis on a negative scale. Would this also be might due to
the attenuator you are using? or should I consider the negative of the
difference?

2) Also you have told to run the UHD calibration routines should be run for
the N210+WBX:

   - uhd_cal_rx_iq_balance
   - uhd_cal_tx_dc_offset
   - uhd_cal_tx_iq_balance

Are these mandatory or optional? Do they affect the daughter board or
mother board of the device? The devices I use are for common use in a lab
and hence it is preferable not to change any of the configurations for
specific use for long term. Hence my question.

As per the information given at
http://files.ettus.com/manual/page_calibration.html#calibration_ it seems
that it more tending towards the WBX board operation. Also the factors get
stored in the machine they are connected to. So in case we want to run the
USRPN210 device with old configuration, it would be ok to just delete the
files or run it from some other machine? Is this correct?

Can any random number be specified for a serial number if not detected from
the device or any specific format is to be used ( i.e min number of digits
& alphabets etc)


Kindly clarify the above points.


Thanks again

Regards

Gayathri


<http://files.ettus.com/uhd_docs/manual/html/calibration.html>


On Sat, Aug 9, 2014 at 9:45 PM, madengr <[email protected]> wrote:

> I updated the repo with results for the N210+WBX:
>
> https://github.com/madengr/usrp_rf_tests
>
> Our single tone sweep results look very similar.  Your IMD plot should be
> in
> dBc; plot the difference in power between one of the test tones and one of
> the IMD3 products.  Mine don't do below -50 dBc since I have the input
> attenuator on the analyzer set for the higher power level.  I need to
> reduce
> it for the lower power levels.
>
> 1) Yes, that would be a calibration factor.  If you measure 0 dBm on the
> spectrum analyzer, and -20 dBFS in the FFT, then 20 dB is the calibration
> factor, so -20 dBFS + 20 dB = 0 dBm.  Note when you change the RX gain,
> then
> the calibration factor changes.  If you increase gain 10 dB, then you must
> decrease the calibration factor 10 dB.
>
> 2) Using the FFT to manually measure level can be a problem if the tone is
> split between bins, so it's better to use a coarse FFT.  Measuring all the
> power in the channel at once may be preferable, and is like using a power
> meter.  Use the complex_to_mag_squared, followed by
> inegrate_with_decimation, then 10*log() +K, where K is your calibration
> factor.  If you decimate down to 1 Hz, i.e. decimation_rate=sample_rate,
> you
> can get very precise power readings.  The accuracy drops off with tone
> power
> due to the wideband noise; just like a power meter.
>
> You would have to look at the USRP FPGA block diagram to find out exactly
> what is going on between the input and the FFT, but it is essentially fine
> tuning with an NCO then many stages of filtering and decimation.  I'm sure
> it affects the amplitude slightly as different filters are used for
> different decimations, and the odd vs. even decimation.
>
> 3)  Compression is the drop in gain (not power).  For example look at this
> table where the USRP TX gain is stepped in 1 dB increments and the output
> power is measured:
>
> USRP_TX_Gain_dB, Pout_dBm,  Gain = Pout_dBm - USRP_RX_Gain_dB
> 3.0, 8.0, 5.0
> 4.0, 9.0, 5.0
> 5.0, 9.9, 4.9
> 6.0, 10.7, 4.7
> 7.0, 11.3, 4.3
> 8.0, 12.0, 4.0 <<-- This is the P1dB
> 9.0, 12.6, 3.6
>
> The P1dB is where the gain has dropped from 5.0 to 4.0.  The P1dB
> referenced
> to the output is 12.0 dBm.  The P1dB referenced to the USRP TX gain setting
> is 8.0 dB.  Notice it is also where the digit after the decimal point
> repeats itself if the input is stepped in 1.0 dB increments; i.e it when
> from 5.zero to 4.zero.  This is the quick n' dirty method of finding P1dB.
> If the input moves in 1 dB steps, all you need to monitor is the most
> significant digit after the decimal point.
>
> 4)  I have not done RX testing.  My signal generators are non-synthesized
> and have no digital interface.  I bought a new one on eBay and it should be
> here next week, but I still need another for a two tone test, not to
> mention
> the components to achieve proper isolation between the two.
>
> Thanks,
> Lou
> KD4HSO
>
>
> Gayathri Ramasubramanian wrote
> > Hi
> >
> > Thanks again fro your explanations. You were correct about the
> > measurements
> > I had sent in earlier. I checked with the power meter and the O/p power
> at
> > 450MHz @ ampl 0.707 and TX Gain of 25 gives 11.5 dBm.
> > I have done the single tone and two-tone tests for USRPN210 +WBX board at
> > 3
> > frequencies of 400 MHz, 900 MHz, 1800 MHz.
> > The settings are similar to yours :AMPL 0.707, TX gain : varied from 0 to
> > 31.
> > Could you please check them and let me know if they are fine.
> >
> > Also I am trying to find the RX side linearity range. For this I
> connected
> > signal generator to the RX port (RF2) of the USRPN210 and viewed the
> > received spectrum on UHD_FFT.
> > I have plotted the readings that I got.
> > I could see that there is a difference of around 35 dBm between the
> > reading
> > on UHD_FFT and that on spectrum Analyser when the same input from signal
> > geenrator was given to both (using a power splitter ) @ 400 MHz.
> > This changed to ~ 31 dBm at 900 MHz and ~ 23 dBm at 1.8 GHz.
> > Could you kindly clarify the below points regd this:
> >
> > 1) So it is good to assume *35 dBm as a calibration factor  @ 400 MHz
> > *for
> > the USRPN210 +WBX device.  i.e in case we use USRPN210 +WBX device test
> as
> > Receiver for testing and use UHD_FFT to plot the spectrum and take the
> > reading of amplitude in dB and get a value of *25 dB* , would it to fine
> > to
> > say that 25 - 35 = -10 dBm is the actual power received by USRP to an
> > extent ( not only on the RX port).
> > So can we use the difference to be the calibration factor in the linear
> > range?
> >
> > 2) The link
> >
> http://www.ettusresearch.com/content/files/kb/application_note_uhd_examples.pdf
> > says that "When the FFT(default) view is used, the x-scale is the
> > frequency, and the y-scale is amplitude.The y-scale shows the amplitude
> > with “counts,” and the values do not typically correlate to a specific,
> > absolute power input. The amplitude read on the display is useful for
> > approximate comparisons. The level for a given input amplitude will vary
> a
> > few dB across frequency and from unit to unit. Also, receiver
> > daughterboards provide various levels of amplification in their analog
> > chains,which will affect the amplitude result in the FFT".
> >
> > Could you clarify what type of processing is done on the received signal
> > from the point of reception till the display on UHD_FFT briefly in terms
> > of
> > scaling/ normalization.
> >
> > The basic idea is to find a relation between the amplitude value shown on
> > UHD_FTT plot and basic power scale in dBm.( a factor to be
> > summed/subtracted or multiplied or divided from UHD_fft reading to get
> > real
> > power value in dBm)
> >
> > 3) On further increase of input power in steps of 1 dBm or 0.1 dBm, there
> > was a 1 dB drop of power. We assumed this was the point of 1 dB
> > Compression
> > for the receiver side of USRPN210. Is this understanding correct? This
> > comes around *-8 dBm @ 400 MHz, -4 dBm @ 900 MHz and ~ +2 dBm @ 1.8 GHz*.
> >
> > 4) I have also tried two tone test on the USRPN210 +WBX device. The plots
> > shown under Q4. The IIP3 point comes ~ 4 dBm when the factor of 35 is
> > subtracted from the values got from UHD_FFT and is ~ 25 dB on IIP3 and 50
> > dBm on OIP3 ( which was told to me by my advisor as not the normal values
> > to expect)  . You said you plan to do the  one tone and two tone test for
> > RX side for B200. Do you find any similarity in results. Kindly let me
> > know
> > if I seem to be going wrong any where.
> >
> >
> > Kindly Clarify the above points.
> > I eagerly look forward to your responses.
> >
> >
> > _______________________________________________
> > Discuss-gnuradio mailing list
>
> > Discuss-gnuradio@
>
> > https://lists.gnu.org/mailman/listinfo/discuss-gnuradio
> >
> >
> > Plots for the Questions.docx (47K)
> > &lt;
> http://gnuradio.4.n7.nabble.com/attachment/49883/0/Plots%20for%20the%20Questions.docx&gt
> ;
>
>
>
>
>
> --
> View this message in context:
> http://gnuradio.4.n7.nabble.com/GR-USRP-and-GPIB-measurements-tp49727p49890.html
> Sent from the GnuRadio mailing list archive at Nabble.com.
>
> _______________________________________________