On Wed, Aug 13, 2014 at 9:59 PM, Paul Stoffregen <[email protected]> wrote:

> A "10X" passive probe attenuates the signal by a factor of 10, but it also
> reduces the load it places on your circuit by a factor of 10.


Only at DC. 10x probes will look like a large part of the scope input
capacitance, say 50% or more. By the same token a 10x probe will have the
full 100 MHz bandwidth, for a 100 MHz scope, whereas the 1x probe will be
15 MHz (rough numbers, you capacitance/bandwidth will vary, etc.). The
capacitive divider of probe + scope doesn't match the 10x ratio of the DC
divider, where the probe is basically a 9 meg series resistor.

It's important that the compensation range of the probe covers a range that
includes the input capacitance of the scope.

If you really want low loading there are active (FET) probes, but they
certainly aren't cheap.

If you want to get a rough idea of how much having the ~15 pF of probe
hooked into your circuit is affecting it simply hook up a second probe,
comparing the waveform before and after. Of course for really sensitive
things like xtal oscillators this might cause it to not work at all. More's
the fun. :-)

Regards,
NealS