Quoting [EMAIL PROTECTED]: > > I found that compared with the initial AGC > threshold setting of 3.80V the level needed to give S9 on the K2 with the > preamp > off had dropped from 63uV PD to 25uV. This was after careful adjustment of > the S LO and S HI required as per the K2 Alignment Instructions
I would recommend that you do this the other way around: adjust AGC setting first for no decrease in level on a quiet band and then the S-meter settings for approximately S9=50 uV. Otherwise you will mix the receiver's sensitivity to quiet bands with S-meter sensitivity. My experience is that the following works well: 1. First set CAL S LO as per the manual 2. Then adjust CAL S HI by setting RF GAIN to the 9 o'clock position (instead of fully CCW) and set parameter so that right-most bargraph is barely turned on. I have played with various other setting such a 10, 11 o'clock, but found the 9 o'clock position to be fine. It also agrees reasonably well with 50 uV from my generator to give S 9. If the AGC-setting is changed, the S-meter calibration has to be redone. Sverre LA3ZA _______________________________________________ Elecraft mailing list Post to: [email protected] You must be a subscriber to post to the list. Subscriber Info (Addr. Change, sub, unsub etc.): http://mailman.qth.net/mailman/listinfo/elecraft Help: http://mailman.qth.net/subscribers.htm Elecraft web page: http://www.elecraft.com