As an electronic designer for laboratory devices, our designs have to 
comply with the IEC 1010 safety standard.
At this moment I have a few questions about this topic.

- Routine tests
  For EMC purposes we use 275 V varistors between fase-earth and 
  neutral-fase. According to annex K.2 of amendement 2, a test voltage 
  of about 2000 Vdc must be applied between the mains terminals 
  connected together (fase and neutral) on one side, and all 
  accessible conductive parts (earth) connected together on the other. 
  No breakdown or repeated flashover shall occur. It's clear that 
  breakdown will occur.
  In annex K.3 it's noticed that if voltage limiting devices are used, 
  the applied voltage may be less. But this exeption does apply for 
  the mains test of annex K.2 to my opinion. Is this correct?

- Interpolation of CREEPAGE DISTANCE
  Annex D.3 of IEC 1010 states that interpolation of CREEPAGE DISTANCE 
  is permissible in accordance with IEC 664 table IV.
  As I'm not in the possession of this standard, I would like to know 
  if this is a liniar interpolation. I will give an example.
  The working voltage is 240 Vac, basic insulation is required, 
  pollution degree 2, and installation (overvoltage) category II. The 
  CREEPAGE DISTANCE is determined in Table D.4.
  
  Working voltage     CREEPAGE distance on printed wiring board
      150V              0.5
      300V              1.5

  Distance = (Working voltage / 150) - 0.5
  For 240 V the CREEPAGE DISTANCE will be 1.1 mm. Is this correct?