Rene,

You are probably doing the "let's get this plastic enclosure to pass EMC".
But, first let's look at any potential safety hazards.  The potential for hazard
is mainly electrical - the coating begins to flake and thereby defeating the 
insulation barrier. 

Now, in terms of regulatory requirements, this is definitely a ULism.  UL has an
approval process to test that both the materials and the process employed are 
suitable to prevent flaking under a range of environmental conditions.  The 
materials and the process are then recognized by UL.

In my experience, we had the coated covers shipped to us with the Condition of 
Compliance in each box or lot. This worked out fine.  The only glitch I remember
was once when UL did not recognize a baby company of the actual recognized 
company.  
 
Dan Levin  
Bioscience Products
HP 

______________________________ Reply Separator _________________________________
Subject: Re: Conductive Coating, IEC 601-1
Author:  Non-HP-owner-emc-pstc ([email protected]) at 
HP-PaloAlto,mimegw3
List-Post: [email protected]
Date:    8/27/96 5:44 AM




On Mon, 26 Aug 1996 [email protected] wrote:

> 
>      
>      
>      Greetings:
>      
>      One of our medical products is being investigated for compliance 
>      to IEC 601-1 based standards ( UL 2601-1, CSA 601.1, and EN 
>      60601-1). We are using some conductive coating for EMI shield. Is 
>      there any requirements on the coating ( i.e. UL 746C..). Do they 
>      care about this in Europe. Is there any US or Canadian deviations 
>      on this ?..
>      
> 
conductive coating on plastic enclosures makes it difficult to recycle 
the plastic material at the end of the life cycle of the equipment, so 
there might be problems if you want to ship to Germany

regards

rene