I have been asked to investigate an issue of Interpolation of CREEPAGE 
AND CLEARANCE regarding the ratings of handheld multimeters.  Clause 
D.2 of IEC 1010-1 (and the harmonized equivalents; ANSI/ISA 
S82.01-1994, CAN/CSA C22.2 No. 1010-1: 1992 and EN 61010-1:1993) 
sets forth  the CREEPAGE and CLEARANCE determination requirements if 
the working voltage is up to 1000 V and if installation categories 
(overvoltage categories) are applicable.  This includes the ratings of most 
handheld DMM's.  Clause D.2.1 points to the applicable tables for BASIC or 
DOUBLE INSULATION and INSTALLATION CATEGORY according to 
annex J.  Clause D2.2 "Application of Tables D.1 to D.12"  allows 
interpolation of CLEARANCE only for a circuit or part which has now 
direct connection to the mains supply...". It further requires CREEPAGE 
DISTANCE to always be at least as large as the value specified for 
CLEARANCE.  

This seems to say that the rating for any product that directly connects to the 
mains supply must be at one of the incremental voltages 
(50,100,150,300,600,1000) listed in Tables D.1 thru D.12.  

I have polled CSA's Vancouver BC office, TUV Product Services in San 
Diego and UL's Primary UL 3111 engineer in Melville.  So far (with the 
exception of UL, whose response is under consideration and review by UL) 
the response has been consistant with the above paragraph.  

Here's the question for the experts at the agencies (UL, TUV, VDE, CSA and 
others), or elsewhere:  Is it possible within the bounds of IEC 1010-1 to 
rate a handheld DMM for example at  750 V and interpolate the CREEPAGE 
and CLEARANCES in spite of the above ?  