Robert,

As I understand the requirement in IEC 601-1 para 20 and following, only 
those parts which provide safety function need to be tested.  My experience 
has been that MOV's are used for the purposes of the EMC directive for 
various surges and line spikes and actually provide not safety function. 
 However the mains input insulation does need to be tested as a routine 
test.  This includes any wire insulation, terminal blocks, circuit breakers, 
fuse holders, EMI filters and so on.  This would be "basic" insulation and 
the test voltages of table 7 row 1 are used.  If the insulation test exceeds 
the clamping value of the MOV then you are allowed to temporarily remove the 
component for the test, but the test still needs to be performed.  In other 
words the protection provided by the MOV's is separate and distinct from the 
insulation being tested.

As a side note be careful using MOV's.  If you are using agency rated MOV's 
such as VDE, then you should be in good shape.  If you are not using rated 
MOV's then there may be trouble.  Especially if your input current 
protection is over 16 Amps.  The MOV's can become a safety problem.  Under 
certain conditions they may catch fire, explode or sputter metal onto nearby 
insulating surfaces.  My safety inspector recommends MOV's protected with 
thermal fuses or placing a spark gap in series with the MOV.

Doug Powell
 -----------------------------
Advanced Energy Industries, Inc.
[email protected]

 ----------
From: Robert Rovai
To: [email protected]
Subject: Insulation test and EMC protections
List-Post: [email protected]
Date: Monday, June 09, 1997 5:30PM


While testing an apparatus (of class I) for EN-60204-1, at the voltage test 
session i meet the 1000V insulation test between the live conductors and the 
protection earth. As the apparatus has an internal electronic circuitry 
protected with MOV'S from line to line and line to ground i ask myself if 
the 1000V withstand test is applicable or not to equipments with this kind 
of protections... or whether such a protection is in contrast with basics 
safety standards.


Robert Rovai
Electronic Department
S.E.Q.W
Pisa - Italy


Reply via email to