My company has been testing our standard power supply products to the ESD test
voltages in IEC601-1-2:1993 (Medical EMC requirements).  The levels are 3kV
contact & 8kV air.

Recently, one of our customers started testing to IEC601-2-24 (Safety of
Infusion Pumps and Controllers).  They said the levels in that spec are 8kV
contact & 15kV air - much higher than our test levels.

Is this a trend for ESD test levels in product-specific standards, of
significantly higher test levels?  Or are other medical product-specific
standards comparable to IEC601-1-2 (ie, with a few kV)?
I realize that the next version of IEC601-1-2 raises the contact discharge
voltage to 6kV.  This doesn't seem like a big change.

If the trend is toward higher voltages in these other standards, I'd like to
find out so we can plan accordingly.

--
Patrick Lawler