Lightbulbs make great AC loads, but use a line filter in series with them
and you'll probably duplicate most EUT's since they'll be using filters.

                                          - Robert -


On Fri, 9 Oct 1998 [email protected] wrote:

> Hi Ed,
> 
> I appreciate your kindness to share your experience with us about 
> constructing equivalent load: "I constructed a load bank consisting of 16 
> surface mount light bulb sockets, all wired in parallel. I just screw in an 
> array of 25/60/75/100 Watt rated lamps until I get the necessary current. 
> Sure, there's some unknown slight lead inductance and capacitance. But all 
> I want to do is draw a few amps DC"
> 
> I have two questions: (1)What is the impedance of your bulb array at 30 
> MHz?   i.e.,  Zb=? @30 MHz.  (2)What is the impedance of EUT at 30 MHz?    
> i.e.,  Ze=? @30 MHz
> 
> If we are not sure Zb=Ze @30 MHz, I am afraid, it's hard to say the 
> spectrum analyzer would receive the same RF emission at 30 MHz from noise 
> sources other than EUT, although the bulb array draws the same current at 
> 60 Hz as EUT does.  In other words, Zb=Ze @60 Hz is one thing, and  Zb=Ze 
> @30 MHz would be another.
> 
> Let's see an example, assuming
>       Ze=Re+jXe, where Xe=Omega*Le, and Omega=2*Pi*F.
>       Re=20 Ohm, Xe=0.1 Ohm @60 Hz,           Ze=20+j*0.1=>20 Ohm
>       Be=20 Ohm, Xe=50000 Ohm @30 MHz,        Ze=>j*50000 Ohm
> 
> Conclusion: As far as the equivalent load is concerned, we can only pay 
> attention to the equivalence of resistance part of Zb and Ze @60 Hz.  At 30 
> MHz, however, we should pay more attention to the equivalence of reactance 
> part of Ze and Zb instead.
> 
> Suggestion: We might need to check the equivalence of Zb and Ze @30 MHz by 
> using an Impedance Analyzer, e. g., HP4191A(?).
> 
> 
> Thank you.
> Please correct me.
> Best Regards,
> Barry Ma


---------
This message is coming from the emc-pstc discussion list.
To cancel your subscription, send mail to [email protected]
with the single line: "unsubscribe emc-pstc" (without the
quotes).  For help, send mail to [email protected],
[email protected], [email protected], or
[email protected] (the list administrators).

Reply via email to