The short answer is: IEC standards are for international use. Those for
european use are EN standards. Therefore, IEC 60601-1-2 is for global
use.

Regards
Vagn Sylvest
DELTA Electronics Testing

>>> Patrick Lawler <[email protected]> 980616-02.16 >>>
IEC601-1-2 (Medical EMC requirements) has a section discussing the
voltage dip and interrupt requirements.  In the Committee Draft dated
1998-03-06, it's section 36.202.5.
It says the test method in IEC61000-4-11 shall apply with
modifications, and goes on to detail them.

What caught my eye was modification 2:
"For equipment and/or systems having multiple voltage settings for
power input, the test shall be performed at EACH RATED INPUT
VOLTAGE.
For equipment and/or systems having, for power input, autoranging
voltage capability, the test shall be performed at the MINIMUM RATED
INPUT VOLTAGE."

(emphasis added by me)

Our power supply products are typical wide-range units, with AC input
voltage specifications of 90-264VAC, and would have little problem
meeting the requirements at 230VAC.  However, meeting voltage dip
and
interrupt requirements at 90VAC would require a redesign.

How did this 'all rated voltages' clause get in the spec?  I was under
the impression that all EMC testing was done at 230VAC, being the
european norm?

--
Patrick Lawler