To me the Medical Device Directive is not the point...........the point
is.........will testing to EN 61000-3-2 make your product safer? If the
answer is "yes" then you should do it. The safety of your customer or an
innocent user is the ultimate "Directive" that must be passed. It is
usually more important to understand the spirit of the law rather than the
letter.

===========================================================================
==============





[email protected] (Jon Griver) on 06/21/99 02:48:15 PM

Please respond to [email protected] (Jon Griver)


To:   "George, David L" <[email protected]>
      [email protected]
cc:    (bcc: Robert E. Heller/US-Corporate/3M/US)
Subject:  EN 61000-3-2 and medical equipment





David,


>I would like to add one remark about some past messages on this discussion
>group.  EN 61000-3-2 contains limits for "professional" equipment.  This
>category applies to medical equipment and any other products used in the
>workplace or in manufacturing activities.
>

I would like to clarify this point about EN 61000-3-2 and medical
equipment.

The scope of EN 61000-3-2 may include medical equipment. This is not the
point. The regulatory requirement in the EU for medical equipment is the
Medical Devices Directive. The only EMC standard which is harmonized for
the MDD is EN 60601-1-2, which does not call out EN 61000-3-2.
See
http://europa.eu.int/comm/dg03/directs/dg3b/newapproa/eurstd/harmstds/refli
s
t/meddevic.html

So, in order to fulfil the regulatory requirements of the EU, medical
equipment does NOT have to comply with EN 61000-3-2.

The inclusion of a product group within the scope of an EN standard does
not necessarily imply that compliance with that standard is mandatory.


Jon Griver

ITL (Product Testing) Ltd
http://www.itl.co.il





---------
This message is coming from the emc-pstc discussion list.
To cancel your subscription, send mail to [email protected]
with the single line: "unsubscribe emc-pstc" (without the
quotes).  For help, send mail to [email protected],
[email protected], [email protected], or
[email protected] (the list administrators).









---------
This message is coming from the emc-pstc discussion list.
To cancel your subscription, send mail to [email protected]
with the single line: "unsubscribe emc-pstc" (without the
quotes).  For help, send mail to [email protected],
[email protected], [email protected], or
[email protected] (the list administrators).

Reply via email to