The FDA does look over the shoulder of pacemaker manufacturers to check
their immunity performance.  They test to AAMI standards and to their own
internal
standards, with many of the tests being performed at levels from 100 to 200
Volts per meter.

TUV Product Service in St. Paul MN does a great deal of pacemaker testing,
possibly testing more pacemakers each year than any other lab.  Also, some
EMC consultants in Minneapolis are very familiar with pacemaker testing
issues.

Paul Cook
NARTE Certified EMC Engineer
Alpha EMC Inc
8540 West River Rd
Minneapolis, Minnestoa 55444
Tel # (612)-561-2844
Fax #(612)-561-3400
E-mail    [email protected]
Specialty  -  EMC Consulting



-----Original Message-----
From: WOODS, RICHARD <[email protected]>
To: IEEE Group <[email protected]>
List-Post: [email protected]
Date: Monday, June 21, 1999 10:26 AM
Subject: RE: Pacemakers


>
>Brian, in the EU all pacemakers must comply with the immunity requirements
>of EN50061/A1:1995. There are no real requirements in the US.
>
>Pacemakers must not malfunction when a 1V peak to peak sine signal with a
>frequency range of 20 Hz-500 Hz is applied to a tissue interface simulation
>circuit connected to the pacemaker.
>
>In another test, the pacemaker must continue to function normally when a
>modulated signal is applied. 2 mVpp, 20 Hz-1 kHz; 2 mVpp X f/1 kHz from 1
>kHz-8 MHz; 16 Vpp, 8 MHz-30MHz.
>
>If you have some real concerns, Georgia Tech Research Institute can test
>your equipment with pacemakers. Contact Jimmy Woody at (404) 894-3535.
> ----------
> From:  Brian At Work [SMTP:[email protected]]
> Sent:  Friday, June 18, 1999 4:01 PM
> To:  IEEE Group
> Subject:  Pacemakers
>
> How do I know if a product MIGHT interfere with a Pacemaker or not?
>What types of emissions and levels are Pacemakers sensitive to?  Does a
>product evaluation check list exist to help determine this.
>
> I know nothing about Pacemakers. BTW, we manufacture laboratory and
>IT equipment.
>
> Thank you,
> Brian
>
>---------
>This message is coming from the emc-pstc discussion list.
>To cancel your subscription, send mail to [email protected]
>with the single line: "unsubscribe emc-pstc" (without the
>quotes).  For help, send mail to [email protected],
>[email protected], [email protected], or
>[email protected] (the list administrators).
>
>
>


---------
This message is coming from the emc-pstc discussion list.
To cancel your subscription, send mail to [email protected]
with the single line: "unsubscribe emc-pstc" (without the
quotes).  For help, send mail to [email protected],
[email protected], [email protected], or
[email protected] (the list administrators).

Reply via email to