Brian, Be sure you are not imposing tests on yourself that are not required. You may protect against accessible parts becoming live under single fault in one of three ways. Try looking at your product as if you are using double insulation (6.5.2) rather than protective bonding. If you can meet the requirements for 6.5.2 then having grounded enclosure panels and the like is just icing on the cake, and not a compliance point you have to test.
Best Regards, Lauren Crane Eaton Corporation > -----Original Message----- > From: Brian At Work [SMTP:[email protected]] > Sent: Friday, September 10, 1999 4:48 PM > To: IEEE Group > Subject: Ground Bond Testers > > Hello Everyone, > > If I have a product that is rated 40 amps at 230V~, I assume I will have > to perform Ground Bond Testing at Twice the Rated current per EN61010-1 > section 6.5.1.2 which would be 80 amps. I have checked by regular sources > and have only found testers that goes to around 30 amps. > > Can anyone send me a source to purchase a Ground Bond Tester that goes up > to maybe 100 amps? > > If such a device does not exist, can anyone give me their recommendations > on how to best perform this test? > > Thank you and best regards, > Brian Kunde --------- This message is coming from the emc-pstc discussion list. To cancel your subscription, send mail to [email protected] with the single line: "unsubscribe emc-pstc" (without the quotes). For help, send mail to [email protected], [email protected], [email protected], or [email protected] (the list administrators).