Hi,

See below for answers to your questions. In summary, IEC 60601-1, Clause
56.11 does not allow this type if construction.

Ned Devine
Entela, Inc.
Program Manager III
Phone 616 248 9671
Fax  616 574 9752
e-mail  [email protected] 


Dear All,

One of the sections in the medical equipment standard (IEC 601-1) 
requires that handheld and foot-operated control devices and their 
cables be in circuits not more than 25 Vac or 60 Vdc or peak value 
separated from the mains by an acceptable method as specified in 
the standard.

Question: Does this requirement apply to a a hand-held applied 
part with push-button switches supplying high voltages to the 
patient?

Answer: Yes.  This type of construction is not allowed.

Question: Are the limits 25 Vac or (60 Vdc or 60 Vpeak)? 

Answer: Yes.  The limits are the SELV limits for medical equipment.  See
Clause 2.4.3.

Question: Are the limits 25 Vac OR 60 Vdc OR any peak value if 
separated from the mains by an acceptable method?

Answer: No.

Question: What is the rationale behind the requirement? Why can't 
someone have a 230 V rated foot-operated device which fully 
complies with the requirements or have a 1000 V applied part 
complying with the requirements?

Answer: Don't know.  My guess is that they did not want to take the chance
of someone inadvertently using a SELV control on a mains circuit.

Regards,




PETER S. MERGUERIAN
MANAGING DIRECTOR
PRODUCT TESTING DIVISION
I.T.L. (PRODUCT TESTING) LTD.
HACHAROSHET 26, P.O.B. 211
OR YEHUDA 60251, ISRAEL

TEL: 972-3-5339022
FAX: 972-3-5339019
E-MAIL: [email protected]
Visit our Website: http://www.itl.co.il

---------
This message is coming from the emc-pstc discussion list.
To cancel your subscription, send mail to [email protected]
with the single line: "unsubscribe emc-pstc" (without the
quotes).  For help, send mail to [email protected],
[email protected], [email protected], or
[email protected] (the list administrators).


---------
This message is coming from the emc-pstc discussion list.
To cancel your subscription, send mail to [email protected]
with the single line: "unsubscribe emc-pstc" (without the
quotes).  For help, send mail to [email protected],
[email protected], [email protected], or
[email protected] (the list administrators).

Reply via email to