Hi All,

ESD or EFT could be the culprit. I would suggest putting a current
probe around one of the instrument cables and connect to a digital
scope and see if you can trigger on a noise event. If you are lucky,
the instrument with the current probe will be the one that goes down.

Alternatively, connect a small antenna to the scope, set to 50 Ohm
input, set the V scale to 100 mV/div, H to 20 ns/div, and trigger on a
50 mV positive going edge for starters. If you get nothing or too
much, adjust the parameters accordingly. A 6 inch diameter loop of
wire makes a good antenna.

Doug

"Price, Ed" wrote:
> 
>         Joe wrote:
> 
>         -----Original Message-----
>         From:   [email protected] [SMTP:[email protected]]
>         Sent:   Tuesday, September 07, 1999 2:04 PM
>         To:     [email protected]
>         Cc:     [email protected]
>         Subject:        Immunity Problems ??
> 
>         >EMC Professionals,
> 
>         >We have 42 of the same instruments installed in one laboratory.
> 
>         What are they? What type of power feeds and I/O lines or sensors?
> What about physical separation?
> 
>         >They are each installed on their own dedicated line.
> 
>         Powerline?
> 
>         >On a regular basis
> 
>         The toughest problems are unpredictable, intermittent failures. This
> is your first ray of hope; you should try to associate the problem with
> external conditions; i.e., a step in your process, the plant elevator, the
> air conditioning, every 107 minutes, something.
> 
>         >an instrument will shut down during a run.
> 
>         A bit more description about the "shutdown."
> 
>         >The instrument that shuts down is different each time.
> 
>         Really, NEVER the same instrument? Or do you just mean it strikes
> randomly?
> 
>         >These instruments are laboratory equipment that meet Class A
> radiated limits of
>         >EN 55011 and meet the immunity requirements of EN 50082-1.
> 
>         So, then you really don't have a problem? (Sorry, grin!)
> 
>         >Any ideas as to the potential cause of the problem and possible
> ways to resolve
>         >the problem would be greatly appreciated.
> 
>         Well, you can start right away with the correlation effort. You
> could also put a powerline monitor on the power buss. If you operated one
> instrument without I/O or data or control leads, you might start to get an
> idea about the problem if that one instrument NEVER shutdown. That might say
> the problem is not coupled through the power bus. OTOH, if it failed just as
> the others, that may lend weight toward the power bus as the path.
> 
>         Can you operate an instrument in another building? Do you get zero
> shutdowns, or the same probability? Can you operate a cluster of your
> instruments in a remote location, and do they exhibit shutdown problems too?
> 
>         >Regards
> 
>         >Joe Martin
>         >EMC/Product Safety Engineer
>         >P.E. Biosystems
>         >[email protected]
> 
>         That's enough for now! Organize your info and get back to the list.
> 
>         Regards,
> 
>         Ed
> 
> :-):-):-):-):-):-):-):-):-):-):-):-):-):-):-):-):-):-):-):-):-):-):-):-):-)
> Ed Price
> [email protected]
> Electromagnetic Compatibility Lab
> Cubic Defense Systems
> San Diego, CA.  USA
> 619-505-2780 (Voice)
> 619-505-1502 (Fax)
> Military & Avionics EMC Services Is Our Specialty
> Shake-Bake-Shock - Metrology - Reliability Analysis
> :-):-):-):-):-):-):-):-):-):-):-):-):-):-):-):-):-):-):-):-):-):-):-):-):-)
> 
> ---------
> This message is coming from the emc-pstc discussion list.
> To cancel your subscription, send mail to [email protected]
> with the single line: "unsubscribe emc-pstc" (without the
> quotes).  For help, send mail to [email protected],
> [email protected], [email protected], or
> [email protected] (the list administrators).

-- 
-----------------------------------------------------------
    ___          _           Doug Smith
     \          / )          P.O. Box 1457
      =========              Los Gatos, CA 95031-1457
   _ / \     / \ _           TEL/FAX: 408-356-4186/358-3799
 /  /\  \ ] /  /\  \         Mobile:  408-858-4528
|  q-----( )  |  o  |        Email:   [email protected]
 \ _ /    ]    \ _ /         Website: http://www.dsmith.org
-----------------------------------------------------------

---------
This message is coming from the emc-pstc discussion list.
To cancel your subscription, send mail to [email protected]
with the single line: "unsubscribe emc-pstc" (without the
quotes).  For help, send mail to [email protected],
[email protected], [email protected], or
[email protected] (the list administrators).

Reply via email to