ANSI/IEEE C95.1 is for human exposure but it does not cover pacemakers.

Draft VDE0848 part 2 (1991) covers 30 kHz-300Ghz and does have limits for
pacemakers. An English translation is available from BSI.

There is no relationship between the limits for human exposure and limits
for pacemakers. The limits for human exposure are based upon induced current
density at the lower frequencies and SAR at the higher frequencies. The
limits for pacemakers are based upon the induced voltage in the lead.

The ANSI limits are as follows:

3kHz-100kHz: 614V/m and 163A/m
100kHz-3MHz: 614V/m and 16.3/f A/m where f is in MHz

the limits are based upon SAR, so the square of values are averaged over any
6 minute period.


Richard Woods

----------
From:  [email protected]
[SMTP:[email protected]]
Sent:  Friday, December 08, 2000 11:22 AM
To:  [email protected]
Subject:  Pacemakers and ANSI standard...?


     What is the ANSI standard or regulation for safe limits for people 
     exposed to electromagnetic fields?  How about people with pacemakers?
     
     I've been informed that there may have been a German study done in 
     1988 concluding that these field limit values could be used in testing 
     the pacemaker immunity.
     
     Can anyone help me to get these ANSI values, at least for the magnetic 
     fields below 400 kHz?
     
     
     David Tarnowski
     Whirlpool Corp.
     
     
     
     
     

-------------------------------------------
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

To cancel your subscription, send mail to:
     [email protected]
with the single line:
     unsubscribe emc-pstc

For help, send mail to the list administrators:
     Jim Bacher:              [email protected]
     Michael Garretson:        [email protected]

For policy questions, send mail to:
     Richard Nute:           [email protected]


-------------------------------------------
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

To cancel your subscription, send mail to:
     [email protected]
with the single line:
     unsubscribe emc-pstc

For help, send mail to the list administrators:
     Jim Bacher:              [email protected]
     Michael Garretson:        [email protected]

For policy questions, send mail to:
     Richard Nute:           [email protected]

Reply via email to