Hi,

No - There is a draft out for IEC 60601-1, 3 edition.  The moved the
Creepage and Clearances section from Clause 57.10 to Section 19.  But, other
than some minor editorial changes, they did not change the requirements. 

Yes - What they did do is remove from the normative references IEC 60664.  

Maybe - They now require a "risk management process".  See Clause 4.6 in the
draft.  I suppose you could use the risk assessment to "prove" that the
spacing you wanted to use were OK.

Ned Devine
Entela, Inc.
Program Manager III
Phone 616 248 9671
Fax  616 574 9752
e-mail  [email protected] 




-----Original Message-----
From: Compliance [mailto:[email protected]]
Sent: Thursday, May 04, 2000 6:14 PM
To: [email protected]
Subject: IEC601 - Spacings Requirements



All -

Is anyone aware of a revision to IEC601 that would affect the creepage and
clearance distance requirements?

Any details or history would be kindly appreciated.

Best Regards,
Brent Taira


-------------------------------------------
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

To cancel your subscription, send mail to:
     [email protected]
with the single line:
     unsubscribe emc-pstc

For help, send mail to the list administrators:
     Jim Bacher:              [email protected]
     Michael Garretson:        [email protected]

For policy questions, send mail to:
     Richard Nute:           [email protected]


-------------------------------------------
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

To cancel your subscription, send mail to:
     [email protected]
with the single line:
     unsubscribe emc-pstc

For help, send mail to the list administrators:
     Jim Bacher:              [email protected]
     Michael Garretson:        [email protected]

For policy questions, send mail to:
     Richard Nute:           [email protected]

Reply via email to