Hello,

1.      Yes.  One of the main goals of the second amendment of IEC/EN
60601-1 was to bring it closer to the Essential Requirements of the Medical
Device Directive.  That is why 60601-1 also references requirements for
biocompatibility and risk analysis.

Also, the EN version of 61010-1 has a requirement for compliance to the EMC
directive.

2.      According to my records, A2 to EN 60601-1 was published in the OJ
(for the MDD) in C 307 on 18 November 1995.  EN 60601-1-2 was published in
the OJ (for the MDD) in C 204 on 09 August 1995.  

I thought that the effective date of the second amendment for IEC 60601-1
was the date of publication (1995).  I would assume that the effective date
for the EN version would also be the date of publication (1995).  

There was a lot of complaints about this, because A2 also required the
remote end of patient leads to be touch proof.  But, the FDA also required
it, so manufactures gave up.  

FYI - The history behind that requirement, was that a nurse in a Chicago
hospital, plugged in the remote ends of an apnea monitor cable, for an
infant, into a IEC 320 detachable power cord.  Seems the lead were colored
Black, White and Green.  If the USA, hospital grade power cords have clear
ends.  The nurse saw the Black, White and Green wires in the IEC 320
connector and plugged in the leads.  The infant died.  I believe that the
FDA had reports of 8 similar incidences.


Ned Devine
Entela, Inc.
Program Manager III
Phone 616 248 9671
Fax  616 574 9752
e-mail  [email protected] 



-----Original Message-----
From: [email protected] [mailto:[email protected]]
Sent: Tuesday, February 29, 2000 5:45 PM
To: EMC-PSTC
Subject: Medical safety & EMC standards tied together?



I recently found out something interesting, and wanted to verify it.

IEC 60601-1 is a standard dealing with electrical safety of medical
equipment, and has two amendments.

IEC 60601-1-2 is a standard dealing with EMC aspects of safety of
medical equipment.

IEC 60601-1 Amendment 2 specifies that paragraph 36 ('Protection
against hazards from unwanted or excessive radiation - Electromagnetic
compatibility') be changed
        From    'Under consideration'
        To      'See IEC 601-1-2'

1) Does this hold for the EN versions as well?  Do statements of
compliance with the basic electrical safety standard automatically
include EMC testing as well?

2) The EN versions were published in the OJ in June 1999.  What is the
DOW of these specifications (if there is one)?


---------
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

To cancel your subscription, send mail to:
     [email protected]
with the single line:
     unsubscribe emc-pstc

For help, send mail to the list adminstrators:
     Jim Bacher:              [email protected], or
     Michael Garretson:    [email protected]
For policy questions, send mail to:
     Richard Nute:           [email protected]


---------
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

To cancel your subscription, send mail to:
     [email protected]
with the single line:
     unsubscribe emc-pstc

For help, send mail to the list adminstrators:
     Jim Bacher:              [email protected], or
     Michael Garretson:    [email protected]
For policy questions, send mail to:
     Richard Nute:           [email protected]

Reply via email to