Jim,

Yes and no.

Yes. Medical ultrasound equipment is explicitly included in the scope of
Part 18 under Section 18.107(f). I assume the logic is that RF circuits are
required to generate the ultrasound.

No. Section 18.121 excludes non-consumer medical ultrasound equipment,
except for a few sections of Part 18 that deal mainly with allowed
frequencies.

Best Regards,

Jon Griver
http://www.601help.com - The Medical Device Developers' Guide to IEC 60601-1


>
> Jon,  Ultrasound equipment does us use RF directly to treat or diagnose
> patients.  I don't believe Ultrasound equipment falls under part 18.
>
> The FDA is now recommending the use of IEC 60601-1-2; 2001 for EMC.
>
> Best regards,
>
> Jim
>
> -----Original Message-----
> From: [email protected]
> [mailto:[email protected]]On Behalf Of Jon Griver
> Sent: Sunday, October 28, 2001 2:07 AM
> To: [email protected]
> Subject: FDA & FCC
>
>
> This is not quite the whole story.
>
> Medical devices must get FDA approval, including EMC aspects. However,
> medical equipment is also subject to FCC regulations.
>
> FCC Part 15 exempts medical equipment (Sec 15.103), though it is still
> subject to the general requirements of the FCC in that devices found to
> cause interference can be stopped from operating.
>
> Medical devices that intentionally use EM radiation are subject
> to FCC Part
> 18. This includes ultrasound equipment, diathermy equipment, microwave
> therapeutic devices.
>
> Regards,
>
> Jon Griver
> http://www.601help.com - The Medical Device Developers' Guide to
> IEC 60601-1
>
>
>
> >
> > Hi,
> >
> > In the USA, it is the FDA.  For most medical products, the FDA
> determines
> > that your product is Substantially Equivalent to a legally
> > marketed device.
> > This is the FDA 510(k) process.  They issue you a letter that
> > allows you to
> > legally market the device.
> >
> > For EMC, the FDA usually wants to see that you comply with IEC
> > 60601-1-2.
> >
> > Ned
> >
> >
> > Ned Devine
> > Program Manager III
> > Entela, Inc.
> > 3033 Madison Ave. SE
> > Grand Rapids, MI  49548
> >
> > 616 248 9671 Phone
> > 616 574 9752 Fax
> > [email protected]  e-mail
> >
> >
> >
> >
> >
> > -----Original Message-----
> > From: [email protected] [mailto:[email protected]]
> > Sent: Friday, October 26, 2001 5:15 AM
> > To: [email protected]
> > Subject: FDA
> >
> >
> >
> > Hi all,
> >
> > What is the basic differences between FDA and FCC ? Don't laugh,
> > yes I know
> > it is a silly question, but if you want to certify medical
> equipment, are
> > the requirements covered in the FDA or in the FCC regulations ?
> >
> > As you understand, within this field, I'm a really novice ...........
> >
> > Best regards
> > Amund Westin, Oslo/Norway
> >
> >
> >
> > -------------------------------------------
> > This message is from the IEEE EMC Society Product Safety
> > Technical Committee emc-pstc discussion list.
> >
> > Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/
> >
> > To cancel your subscription, send mail to:
> >      [email protected]
> > with the single line:
> >      unsubscribe emc-pstc
> >
> > For help, send mail to the list administrators:
> >      Michael Garretson:        [email protected]
> >      Dave Heald                [email protected]
> >
> > For policy questions, send mail to:
> >      Richard Nute:           [email protected]
> >      Jim Bacher:             [email protected]
> >
> > All emc-pstc postings are archived and searchable on the web at:
> >     No longer online until our new server is brought online and the old
> > messages are imported into the new server.
> >
> > -------------------------------------------
> > This message is from the IEEE EMC Society Product Safety
> > Technical Committee emc-pstc discussion list.
> >
> > Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/
> >
> > To cancel your subscription, send mail to:
> >      [email protected]
> > with the single line:
> >      unsubscribe emc-pstc
> >
> > For help, send mail to the list administrators:
> >      Michael Garretson:        [email protected]
> >      Dave Heald                [email protected]
> >
> > For policy questions, send mail to:
> >      Richard Nute:           [email protected]
> >      Jim Bacher:             [email protected]
> >
> > All emc-pstc postings are archived and searchable on the web at:
> >     No longer online until our new server is brought online and
> > the old messages are imported into the new server.
> >
>
>
> -------------------------------------------
> This message is from the IEEE EMC Society Product Safety
> Technical Committee emc-pstc discussion list.
>
> Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/
>
> To cancel your subscription, send mail to:
>      [email protected]
> with the single line:
>      unsubscribe emc-pstc
>
> For help, send mail to the list administrators:
>      Michael Garretson:        [email protected]
>      Dave Heald                [email protected]
>
> For policy questions, send mail to:
>      Richard Nute:           [email protected]
>      Jim Bacher:             [email protected]
>
> All emc-pstc postings are archived and searchable on the web at:
>     No longer online until our new server is brought online and the old
> messages are imported into the new server.
>
>
> -------------------------------------------
> This message is from the IEEE EMC Society Product Safety
> Technical Committee emc-pstc discussion list.
>
> Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/
>
> To cancel your subscription, send mail to:
>      [email protected]
> with the single line:
>      unsubscribe emc-pstc
>
> For help, send mail to the list administrators:
>      Michael Garretson:        [email protected]
>      Dave Heald                [email protected]
>
> For policy questions, send mail to:
>      Richard Nute:           [email protected]
>      Jim Bacher:             [email protected]
>
> All emc-pstc postings are archived and searchable on the web at:
>     No longer online until our new server is brought online and
> the old messages are imported into the new server.
>


-------------------------------------------
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/

To cancel your subscription, send mail to:
     [email protected]
with the single line:
     unsubscribe emc-pstc

For help, send mail to the list administrators:
     Michael Garretson:        [email protected]
     Dave Heald                [email protected]

For policy questions, send mail to:
     Richard Nute:           [email protected]
     Jim Bacher:             [email protected]

All emc-pstc postings are archived and searchable on the web at:
    No longer online until our new server is brought online and the old 
messages are imported into the new server.

Reply via email to