I read in !emc-pstc that Paolo Peruzzi <[email protected]> wrote
(in <[email protected]>) about 'Medical
devices radiated immunity', on Thu, 6 Dec 2001:

>The second edition of IEC 60601-1-2 (EMC for medical electrical equipments)
>sets out a radiated immunity test between 80 MHz and 2.5 GHz. For test
>methods it refers to IEC 61000-4-3, which covers frequencies up to 2 GHz.
>So what about the test method? Can you follow the procedure described in
>Annex J of new 61000-3-4
61000-4-3
> ( FDIS version) as "indipendent windows method" up
>to 2.5 GHz?

Not until the FDIS has passed its vote (closing date 2002-02-01), and
not in Europe until the new edition has been notified in the Official
Journal.

But since there is an obvious error in 60601-1-2, in calling up
61000-4-3 for measurements above 2 GHz, which it does not cover, I think
you can make a case in your technical file for using the method.
-- 
Regards, John Woodgate, OOO - Own Opinions Only. http://www.jmwa.demon.co.uk 
After swimming across the Hellespont, I felt like a Hero. 

-------------------------------------------
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/

To cancel your subscription, send mail to:
     [email protected]
with the single line:
     unsubscribe emc-pstc

For help, send mail to the list administrators:
     Michael Garretson:        [email protected]
     Dave Heald                [email protected]

For policy questions, send mail to:
     Richard Nute:           [email protected]
     Jim Bacher:             [email protected]

All emc-pstc postings are archived and searchable on the web at:
    No longer online until our new server is brought online and the old 
messages are imported into the new server.

Reply via email to