I agree with Ed that there is no problem using switches, but I don't see the
need for 18 GHz, all the measurements you describe stop at 30 MHz.  I don't
see why Mini-Circuits would be out-of-line here, and you can check losses by
using a tracking generator or known signal sources.

----------
>From: "Price, Ed" <[email protected]>
>To: [email protected]
>Subject: RE: Coaxial Switches - use with spectrum analyzer and gear?
>Date: Thu, May 23, 2002, 11:53 AM

>>-----Original Message-----
>>From: Jim Eichner [mailto:[email protected]]
>>Sent: Wednesday, May 22, 2002 12:55 PM
>>To: [email protected]
>>Subject: RE: Coaxial Switches - use with spectrum analyzer and gear?
>>
>>
>>
>>In our pre-compliance lab, we've got a pair of LISN's, a LISN-mate type
>>thing that I'll call a DMRN (DM rejection network), a
>>transient limiter,
>>some clamp-on RF current probes, an amplifier, the front end of the
>>analyzer, and a 50 ohm terminator.  All this gear is connected and
>>unconnected a hundred times a day when we're deep into
>>trouble-shooting, and
>>I'm sick of doing up and undoing coax connectors.  I don't
>>trust push-on
>>coax connectors.
>>
>>What I envision is a system with a few of these coax switches
>>in it. The
>>analyzer input would go to a switch that selected either the
>>amp (used only
>>with the clamp-on probes) or the transient limiter (used with
>>the LISN's).
>>The limiter input would go to a switch that selected between
>>the output of
>>the DMRN or the Line LISN or the Neutral LISN.  Etc.  You get the idea.
>>
>>Is there anything to stop me from using coax switches with suitable CW
>>power, insertion loss, and frequency spec's? Am I headed for
>>trouble if I
>>have 2 or 3 of these switches in the signal path, due to cumulative
>>insertion loss?
>>
>>Thanks for your feedback,
>>
>>Regards,
>>Jim Eichner, P.Eng.
>>Manager, Engineering Services
>>Xantrex Technology Inc.
>>Mobile Power
>>web: www.xantrex.com <http://www.xantrex.com>
>>Any opinions expressed are those of my invisible friend, who really
>>exists. Honest.
>>
>
>
> Jim:
>
> There's no reason why you can't implement the switching system you
> described. The only concern is that you use really good parts. (Your
> spectrum analyzer likely has a couple of internal relays in its signal
> path.) Since you likely need to measure up to about 5 GHz, I would use
> components rated for up to 18 GHz. This is not the place to skimp on cost.
> Don't use Dow-Key relays with SO239 UHF connectors; go with something like
> Agilent relays with SMA connectors. Connect the internal paths with
> semi-rigid solid-wall coax, and use a good grade of flexible external coax
> cabling too. If your system is computer-controlled, you can use something
> like an IEEE488 Relay Driver interface to automate the switching functions.
>
> Your lab procedures should be amended to include some type of periodic
> verification of the loss along all signal paths.
>
> Regards,
>
> Ed
>
> Ed Price
> [email protected]
> Electromagnetic Compatibility Lab
> Cubic Defense Systems
> San Diego, CA  USA
> 858-505-2780  (Voice)
> 858-505-1583  (Fax)
> Military & Avionics EMC Services Is Our Specialty
> Shake-Bake-Shock - Metrology - Reliability Analysis
>
> -------------------------------------------
> This message is from the IEEE EMC Society Product Safety
> Technical Committee emc-pstc discussion list.
>
> Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/
>
> To cancel your subscription, send mail to:
>      [email protected]
> with the single line:
>      unsubscribe emc-pstc
>
> For help, send mail to the list administrators:
>      Ron Pickard:              [email protected]
>      Dave Heald:               [email protected]
>
> For policy questions, send mail to:
>      Richard Nute:           [email protected]
>      Jim Bacher:             [email protected]
>
> All emc-pstc postings are archived and searchable on the web at:
>     http://ieeepstc.mindcruiser.com/
>     Click on "browse" and then "emc-pstc mailing list"
> 

-------------------------------------------
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/

To cancel your subscription, send mail to:
     [email protected]
with the single line:
     unsubscribe emc-pstc

For help, send mail to the list administrators:
     Ron Pickard:              [email protected]
     Dave Heald:               [email protected]

For policy questions, send mail to:
     Richard Nute:           [email protected]
     Jim Bacher:             [email protected]

All emc-pstc postings are archived and searchable on the web at:
    http://ieeepstc.mindcruiser.com/
    Click on "browse" and then "emc-pstc mailing list"

Reply via email to