I take your point Richard, but I did not make mine very clearly and so you have missed it.

Irrespective of the wording of some of the Directives and the guidance, since there is no absolute requirement (except in the medical devices directives) to appoint an Authorised Representative within the EU, any 'requirement' to keep the documentation on EU soil is both nonsensical and unenforceable. If the manufacturer is outside the Community and there is no Authorised Representative, who is going to hold the file, and how will any enforcement action be taken?

I know the pat answer is 'the person responsible for importing the product into the EU' but what if they are also outside the EU? In many cases the only person located within the EU is the end user and you can't seriously tell me that the authorities expect that every time a manufacturer ships to a new end user in the EU they are going to have to supply a copy of their Technical File with the goods!

The Directives are in a bit of a mess on this topic but frankly it probably isn't in anyone's interests to sort it out since the current fudge allows everyone involved to do their own thing without having to obviously break any rules. I can't see the WTO liking a move by the Europeans to force non-EU manufacturers to formally appoint an Authorised Representative, even if such a move were the answer to the problem, and actually (for reasons which are explained on my web site at http://www.conformance.co.uk/CE_MARKING/reps.html) I don't actually think it is the answer anyway.

Regards

Nick.

At 10:23 am +0000 11/12/02, Richard Hughes wrote:
Nick,

Unfortunately, regarding the last para, you have not followed your own good advice of checking with the specific EU Directives concerned.
Annex IV section 2 of the LVD requires:

"The manufacturer must establish the technical documentation described in point 3 and he or his authorized representative established within the Community must keep it on Community territory at the disposal of the relevant national authorities for inspection purposes for a period ending at least 10 years after the last product has been manufactured."

Not surprisingly, the Commission's Guide to the LVD (2001 edition) states:

"This technical documentation must be kept within the Community."


On the other hand, the R&TTED does not state that the technical documentation must be kept within the Community. The corresponding para to the above (Annex II section 2) requires that:

"The manufacturer must establish the technical documentation described in point 4 and he or his authorised representative established within the Community must keep it for a period ending at least 10 years after the last product has been manufactured at the disposal of the relevant national authorities of any Member State for inspection purposes."

Note that the part requiring that the technical documentation be kept "on Community territory" is absent.


Equally, the LVD and the R&TTED have different requirements regarding the provision of a DoC. See Article 6(3) of the R&TTED.


Own opinions as always,

Richard Hughes



-----Original Message-----
From: Nick Williams [<mailto:[email protected]>mailto:[email protected]]
Sent: 11 December 2002 00:12
To: Stephen Irving
Cc: [email protected]
Subject: Re: Location of CE DoCs

These are indeed rumors. Different Directives have different
requirements in this regard, and although the general rules of thumb
that documentation should be kept in the EU and a copy of the DofC
should be supplied with the product provide a good working basis for
compliance with the Directives, they are by no means mandatory for
all equipment under all New Approach directives. If it really matters
to you, you should read the text of the specific directives which
apply to your products in order to find out what is required.

As examples, neither the LVD nor the EMC Directives require a copy of
the EC declaration to be shipped with the product, although the
Machinery Directive does. None of these three directives require the
appointment of an Authorised Representative (only the medical devices
directives do this) and ergo there is no requirement for the
technical file to be 'kept on EU soil'.

Regards

Nick.


-------------------------------------------
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/

To cancel your subscription, send mail to:
    [email protected]
with the single line:
    unsubscribe emc-pstc

For help, send mail to the list administrators:
    Ron Pickard:              [email protected]
    Dave Heald:               [email protected]

For policy questions, send mail to:
    Richard Nute:           [email protected]
    Jim Bacher:             [email protected]

All emc-pstc postings are archived and searchable on the web at:
   http://ieeepstc.mindcruiser.com/
   Click on "browse" and then "emc-pstc mailing list"

Reply via email to