John,

In the United States, the FDA regulates medical devices. The following link 
will gives you information on the FDA's approach to EMC:

http://www.fda.gov/cdrh/emc/

Although IEC 60601-1-2 is a 'concensus standard', FDA investigators sometimes 
have different or additional requirements. I have even 
known cases of demands for compliance with military immunity standards. The FDA 
writes the following:

"This standard should serve as a baseline for all of the appropriate device 
areas. However, where there are existing CDRH guidance 
documents with additional EMC specifications then these will supersede the 
basic IEC requirements." 

Note that you will need UL 2601-1, but for product safety, not EMC. The 
reference to electrostatic charges in clause 39.3, which you 
found, is in regard to the risk of explosion of anaesthetic gases! This 
standard, or the international version, IEC 60601-1, covers 
electrical, mechanical and other safety aspects of medical electrical products.

You may also find EMC requirements in the standards for particular types of 
medical equipment. These are numbered IEC 60601-2-
XX, and you can see a list of them at:

http://www.601help.com/Other_601_Standards/other_601_standards.html

These EMC requirements may amend the requirements in IEC 60601-1-2 and will 
always be in clause 36 of these standards.

Also note that in the US, medical equipment which intentionally uses EM 
radiation (e.g. microwave therapy devices), as well as 
ultrasound equipment, is required to comply with FCC Part 18, for ISM equipment.


For the European Union, the EMC requirements are EN 60601-1-2, plus any EMC 
requirements of EN 60601-2-XX. (The EN version is 
usually almost identical to the IEC version). However, you should note that at 
the moment the 1993 version of the standard is in 
effect, not the 2001 version! Presumably the 2001 version will come into effect 
at some point in the future. You can see a list of all the 
Medical Device Directive harmonised standards at:

http://europa.eu.int/comm/enterprise/newapproach/standardization/harmstds/reflist/meddevic.html

(If this wraps to two lines, you may have to enter the second line manually)

A further note about IEC or EN 60601-1-2. It is a 'pointing' standard, in that 
it refers to and provides test levels for EN 55011 for 
emissions and IEC 61000-4-X for immunity. These latter standards basically 
provide test methods for different EMC phenomena.


Regards,


Jon Griver
www.601help.com
The Medical Device Designer's Guide to IEC 60601-1

********************************************************************

EMC-PSTC'ers,
Does the United States have any electromagnetic compatibility (EMC),
electromagnetic interference (EMI), or electrostatic discharge (ESD)
standards for medical electronics?  If so, where can I find them?

A prospective client has asked about EMC/EMI/ESD testing for a wearable
medical-monitor.  It looks like the applicable standards for Europe are:
*  EN 55011:1998 +Amendment A1:1999.
*  EN 60601-1-2:2001.

I found some references on the Internet to:
*  The Food and Drug Administration using UL 2601-1 for medical 
   devices.
*  UL 2601-1 having the same immunity requirements as EN 60601-1-2.

Having been bitten by supposedly-identical standards being just a little
different in areas critical to products I've designed, we bought
UL2601-1 ($155 with shipping and handling), and received it yesterday. 
The cover letter is dated June 16,2000, and describes it as "Second
Edition, Dated October 24, 1997".

Section 39.2 has a couple of paragraphs about "Prevention of
electrostatic charges".  

Appendix A1, section A1.5.2, says "The sensitivity of EQUIPMENT to
external interference (electromagnetic field, perturbations of the
supply voltage) is under consideration.

Appendix L refers to IEC 601-1-2:1993.

Is there a newer amendment to UL 2601-1 that has EMC/EMI/ESD
requirements, or just what *do* people use to verify/certify medical
electronics?

Thanks!
                                        John Barnes
                                        dBi Corporation
                                        http://www.dbicorporation.com/

-------------------------------------------
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/

To cancel your subscription, send mail to:
     [email protected]
with the single line:
     unsubscribe emc-pstc

For help, send mail to the list administrators:
     Ron Pickard:              [email protected]
     Dave Heald:               [email protected]

For policy questions, send mail to:
     Richard Nute:           [email protected]
     Jim Bacher:             [email protected]

All emc-pstc postings are archived and searchable on the web at:
    http://ieeepstc.mindcruiser.com/
    Click on "browse" and then "emc-pstc mailing list"

------- End of forwarded message -------

-------------------------------------------
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/

To cancel your subscription, send mail to:
     [email protected]
with the single line:
     unsubscribe emc-pstc

For help, send mail to the list administrators:
     Ron Pickard:              [email protected]
     Dave Heald:               [email protected]

For policy questions, send mail to:
     Richard Nute:           [email protected]
     Jim Bacher:             [email protected]

All emc-pstc postings are archived and searchable on the web at:
    http://ieeepstc.mindcruiser.com/
    Click on "browse" and then "emc-pstc mailing list"

Reply via email to