Assuming the sensed current in the coil is reasonably sinusoidal, skin
depth at 100KHZ is about half the diameter of AWG 28, (if the winding is
only one layer thick). This means that the entire wire cross sectional
area will be carrying current. This was my reason for suggestion that
particular wire size. 
 
Actually, given the very small level of induced current (probably
microamps or less), it really makes no difference at all what wire is
used. Even AWG 14 will do the job nicely since even though only the skin
of the larger size wire will be carrying current, there will still be
far more actual effective conducting area than is needed for such a
miniscule current. It's not like this is a high power switchmode power
supply transformer, or something like that, where current density in the
wire is high and minimizing eddy current losses is critical.
 
Bob Wilson 
TIR Systems Ltd. 
Vancouver. 
-----Original Message-----
From: Ken Javor [mailto:[email protected]] 
Sent: June 20, 2002 6:31 PM
To: Robert Wilson; [email protected]
Subject: Re: magnetic field measuring instruments and probes
 
I was simply reporting the military standard construction.  I haven't
done the analysis to determine skin effects and whether using Litz wire
to counter it is useful.

----------
From: "Robert Wilson" <[email protected]>
To: <[email protected]>
Subject: RE: magnetic field measuring instruments and probes
List-Post: [email protected]
Date: Thu, Jun 20, 2002, 5:06 PM


Litz wire at 100KHz sounds a little over the top. AWG20 or 30 will do
the job nicely. 
Bob Wilson 
TIR Systems Ltd. 
Vancouver. 
-----Original Message-----
From: Ken Javor [mailto:[email protected]] 
Sent: June 20, 2002 7:11 AM
To: Istvan Novak; [email protected]
Cc: Istvan Novak - Board Design Technology
Subject: Re: magnetic field measuring instruments and probes 

 

At the low end you could use a 13 cm loop such as that used for military
RE01/ RE101 testing (36 turns of 7-41 Litz wire, electrostatically
shielded).  That loop can be used from 30 Hz to at least 100 kHz.  It is
handy because it is small and can be used as a probe.  There are
numerous loops developed for VDE testing that operate from 10 kHz to 30
MHz.  But these loops are physically large and developed for use at 3
meters from the test item.  There are also close field probes for
diagnostic testing, but these have large antenna factors.  So the answer
depends on your specific need.  

----------
From: "Istvan Novak" <[email protected]>
To: <[email protected]>
Cc: "Istvan Novak - Board Design Technology"
<[email protected]>
Subject: magnetic field measuring instruments and probes
List-Post: [email protected]
Date: Thu, Jun 20, 2002, 7:28 AM


Hi list members,
 
Could someone suggest or refer me to pointers/manufacturers of probes
and instruments to measure magnetic close field in the 10kHz to 10MHz
frequency range?  High accuracy is not important, values to be measured
would be around 0.5mT and up.
 
Thanks
 
Istvan Novak
SUN Microsystems 

Reply via email to