I read in !emc-pstc that Jon Griver <[email protected]> wrote (in
<3E5C9943.14190.33FECD@localhost>) about 'D of C for medical P.S.' on
Wed, 26 Feb 2003:
>You are right not issuing a D of C to the MDD, as your product is not a
medical 
>device or an 
>accessory ot a medical device.
>
>On the other hand I see no reason not to issue a D of C to the LVD.
Compliance 
>with a standard in 
>the list of Harmonised Standards is not the only way to demonstrate
compliance 
>with the LVD. It 
>does not take very much effort to draw up a comparison table showing that the 
>requirements for a 
>power supply in IEC 60601-1 are more stringent than say, IEC 60950 or IEC 
>61010-1. Using 
>engineering judgment on this basis, I believe that you can issue a D of C to
the 
>LVD, specifying 
>IEC 60601-1 as the applied standard. You can present your comparison table to 
>anyone who questions 
>this, demonstrating the engineering basis for your declaration.

That's what I would recommend, but I don't know enough about 60601-1 to
be definite.
-- 
Regards, John Woodgate, OOO - Own Opinions Only. http://www.jmwa.demon.co.uk 
Interested in professional sound reinforcement and distribution? Then go to 
http://www.isce.org.uk
PLEASE do NOT copy news posts to me by E-MAIL!


This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/

To cancel your subscription, send mail to:
     [email protected]
with the single line:
     unsubscribe emc-pstc

For help, send mail to the list administrators:
     Ron Pickard:              [email protected]
     Dave Heald:               [email protected]

For policy questions, send mail to:
     Richard Nute:           [email protected]
     Jim Bacher:             [email protected]

Archive is being moved, we will announce when it is back on-line.
All emc-pstc postings are archived and searchable on the web at:
    http://www.ieeecommunities.org/emc-pstc

Reply via email to