I currently have a conformally coated board that I need to be tested according
to EN 60730-1. The certification agency requires that my conformal coat be a
Type A conformal coat according to EN 60730-1. (In other words it must fufill
annex P and Annex Q of EN 60730-1)
I currently am using a Humiseal 1B73LOC conformal coat on the board. The
agency requires that I have documentation that sates the coating meets the
requirements for EN60730-1. I have called the manufacturer and they do not
seem to know anything about this requirement. Whomever I speak with regarding
this requirement, does not seem to know anything about this. Can anyone give
me a supplier name that is familiar with this requirement? If you have any
experience with this issue can you please let me know what you did to solve
this issue?
Regards and thank you in advance for your help.
Regards,
Paul S. Denomme
Design Engineer
Viasystems
804-226-5155
804-226-5193(fax)
From: [email protected] [mailto:[email protected]]
Sent: Thursday, February 20, 2003 6:27 AM
To: [email protected]
Subject: RE: IEC 61010 requirements
Mr Poore
While you are correct in the intent and meaning of the standard; my point was,
that if the customer specs product compliance with standard xyz, than the
requirements of xyz becomes your design and test target. I have lost count of
the number of times that a customer has spec'd 601-1 or whatever, when either
x950 or x1010-1 is applicable for the intended end use. What the customer
wants, the customer gets (if he pays)...
Brian
There are no requirements for using an isolation transformer as long as you
can meet the requirements for "touch" current (leakage current) without one.
These requirements are similiar to 950, and not as stringent as the medical
requirement.
Chris Poore
-----Original Message-----
From: [email protected]
Sent: Wednesday, February 19, 2003 11:42 AM
To: [email protected]
Subject: RE: IEC 61010 requirements
If end-use wiil never see contact w/patient or test subject, and equipment can
be classified as Electrical Test and/or Measurement equipment, and does NOT
fall under the scope of the Med Directive, and can be scoped under the LVD,
then 61010-1 is applicable standard.
Of course, the customer spec can require compliance to the Med Dev Dir.
Brian
-----Original Message-----
From: Han, Delphina
Sent: Wednesday, February 19, 2003 9:10 AM
To: [email protected]
Subject: IEC 61010 requirements
Hi
I am trying to find out safety requirements for devices that control and
monitor equipment in a hospital lab (used for pathology). Does it fall under
the IEC 61010 standard? If so, are there any requirements for use of
isolation transformers in that standard?