Ken,

Not knowing what medical device you are talking about,
lets start with IEC 60601-1-2 2nd edition 2001-09 the
"medical electrical equipment" umbrella document.  

For radiated RF electromagnetic fields (section 36.202.3, 
general requirements) it calls out 3 V/m from 80 Mhz to 2.5 Ghz
@ 1 Khz AM modulation.  Note, if you are life support it's 
10 V/m @ 2 Hz AM modulation.

Note, IEC 60601-1-2 (section 36.202.6) also requires RF 
conducted disturbances testing, (61000-4-6,) 150 Khz to 80 Mhz
@ 3 V/m.  Note, life support equipment gets 10 V/m treatment
at the ISM freqs within this range.

In a past life (with Ultrasound) RF conducted immunity is
an interesting test.  Contact me for further details.

Regards,

dave garnier

                        David Garnier
                        e GE Health Care
                        ___________________________________________
                        David S. Garnier
                        Senior Technician
                        Functional & CT Engineering
                        3000 N. Grandview Ave - M/S W-1250
                        Waukesha, Wi. 53188
                        Tel: 262.312.7246
                        Cel:  414-899-7580 (NEW)





From: [email protected]
[mailto:[email protected]]On Behalf Of Ken Javor
Sent: Monday, May 10, 2004 5:41 PM
To: [email protected]
Subject: medical device radiated immunity frequency range



I assume the medical device product category includes EN61000-4-3 as a
requirement.  For medical devices, is EN61000-4-3 extended down to 26
MHz,
or is it the usual 80-1000 MHz?

Thank you.



This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

Visit our web site at:  http://www.ieee-pses.org/

To cancel your subscription, send mail to:
     [email protected]
with the single line:
     unsubscribe emc-pstc

For help, send mail to the list administrators:
     Ron Pickard:              [email protected]
     Dave Heald:               [email protected]

For policy questions, send mail to:
     Richard Nute:           [email protected]
     Jim Bacher:             [email protected]

All emc-pstc postings are archived and searchable on the web at:
    http://www.ieeecommunities.org/emc-pstc


This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

Visit our web site at:  http://www.ieee-pses.org/

To cancel your subscription, send mail to:
     [email protected]
with the single line:
     unsubscribe emc-pstc

For help, send mail to the list administrators:
     Ron Pickard:              [email protected]
     Dave Heald:               [email protected]

For policy questions, send mail to:
     Richard Nute:           [email protected]
     Jim Bacher:             [email protected]

All emc-pstc postings are archived and searchable on the web at:
    http://www.ieeecommunities.org/emc-pstc

Reply via email to