It's an Applied Part if it could make contact with the patient during normal
use of the device (whether or not such contact is needed for the device to
fulfil its intended function).
Impedance to earth is not a consideration in determining whether it's an
Applied Part or not.
You could only consider it as Enclosure if it won't make patient contact
during use. Many articulated-arm medical laser systems commonly will make
such contact - whether your system could you will need to judge from its
intended application.

If on this basis your arm is an Applied Part, then it can't be Type BF or CF
if it has low impedance to earth - it would have to be Type-B. Although the
generic standard EN-60601-1 allows  use of Type-B Applied Parts, some of the
Particular Standards for medical devices do not (typically depending upon
the nature of the device and its intended use).


Ged Dean




From: standards [mailto:[email protected]]
Sent: 14 July 2004 16:27
To: [email protected]
Subject: medical laser equipment applied parts


http://www.ieee-pses.org/symposium
          http://www.emc2004.org/


Dear all,
Let's consider a medical laser device with delivers its output (laser beam)
through an articulated mechanical arm; the arm is metallic and connected to
the chassis. The arm has no electrical connections to the rest of the
equipment.
I'm in trouble with leakage currents and applied voltage testing.
My question is: according to EN 60601-1, if the impedance from the arm to
earth is less than 100 mohms, can I think of it simply as an extension of
the
chassis, instead of an applied part? And if the impedance is more than 100
mohms?

thanks,
Paolo


Paolo Peruzzi
Regulatory Engineer
El.En. S.p.A.
Via Baldanzese, 17
50041 Calenzano (FI)
Italy

Tel. +39 055 8826807
FAX  +39 055 8832884
http://www.elengroup.com/



This message is from the IEEE Product Safety Engineering Society
emc-pstc discussion list.

IEEE PSES Main Website:  http://www.ieee-pses.org/

To post a message send your e-mail to [email protected]

Instructions for use of the list server:

    http://listserv.ieee.org/listserv/request/user-guide.html

List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:

     Ron Pickard:              [email protected]
     Dave Heald:               [email protected]

For policy questions, send mail to:

     Richard Nute:           [email protected]
     Jim Bacher:             [email protected]

All emc-pstc postings are archived and searchable on the web at:

    http://www.ieeecommunities.org/emc-pstc



This message is from the IEEE Product Safety Engineering Society
emc-pstc discussion list.

IEEE PSES Main Website:  http://www.ieee-pses.org/

To post a message send your e-mail to [email protected]

Instructions for use of the list server:

    http://listserv.ieee.org/listserv/request/user-guide.html

List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:

     Ron Pickard:              [email protected]
     Dave Heald:               [email protected]

For policy questions, send mail to:

     Richard Nute:           [email protected]
     Jim Bacher:             [email protected]

All emc-pstc postings are archived and searchable on the web at:

    http://www.ieeecommunities.org/emc-pstc

Reply via email to