I have watched this subject unfold with interest.  However it seems the one
key element at the heart of this interchange has been left out of the
discussion.  That being Uncertainty of measurement.  
We are assuming that the lab in question did perform signal maximization of
the cables and did set up the EUT correctly.
The purpose of CISPR 16-4 (2002) " Uncertainty in EMC Measurements",  was to
help bring the issue of measurement deviations between different measurement
sites into some kind of relational understanding.  In Paragraph 4.1 it states:
"...... the expanded measurement uncertainty for the test lab shall be stated
in the test report"
This is the only way a customer has to evaluate whatever margin to limit
exists with that particular product.
To the best of my understanding most labs do not state their uncertainty of
measurement for reasons best known to them.
When I send out a product that I can't do in my lab I insist on adherence to
CISPR 16-4. If the uncertainty of measurement is not included in the report
its of little value to me.
 
If more customers insisted on this information (backed up by real numbers in
the calculations) these sort of problems would resolve themselves in short
order. 
 
Just one guys opinion.
Michael Taylor - NCE
Loveland, CO

  _____  

From: [email protected] 
mailto:[email protected]] On Behalf Of Geetha Balasubramanian
Sent: Tuesday, March 22, 2005 8:52 AM
To: Grasso, Charles; 'Bob Richards '; '[email protected] ';
'[email protected] '
Subject: RE: Radiated Emission Testing


Hi Gurus,
 
           As per the US lab report, the frequency at which the device had
minimum pass margin is 68 MHz. At 68 MHz in vertical polarisation the margin
is 4.7 dB & in horizontal polarisation the margin is 1.6 dB. Measurable
emissions were found only in 40 , 43, 52.38, 64 & 68 MHz.
          But as per the Taiwan lab report no measurable emission is found
>from the device at the above frequencies. The device failed in 122.61 MHz at
horizontal polarisation. 
          There are 2 clock sources in the device, one runs at 12 MHz and the
other runs at 4 MHz. Frequencies showing up in the US lab report are atleast
the harmonics of these 2 frequencies. But frequencies mentioned in the Taiwan
lab report are looking funny (111.27, 135.3, 285.69 MHz. etc.)
           US lab tested in OATS but the TAIWAN lab tested in a anechoic
chamber. I heard that chambers are not the correct way to measure ( always the
results will be wrong) that is why FCC never accepts chamber measurement. Is
it so? 
            Thanking you all for the time & efforts spent on this.
 
Thanking you all in anticipation
 
Geetha

"Grasso, Charles" <[email protected]> wrote:

Hi Bob,

Don't you ALWAYS maximize cables. I found that
with cable maximization my test repeatabilty
improved dramatically.

HAs cable maximization fallen out of favour?



From: [email protected]
To: [email protected]
Sent: 3/22/2005 6:58 AM
Subject: Re: Radiated Emission Testing

It is not an unusual occurance. This is one good
reason why you want a good margin under the limit.
Most people like to see a margin of at least 6dB.

One variable is the site itself. A good OATS is the
best. Sites can have up to 4dB variation from
theoretical, and one chamber I've been involved with
had a problem in the 50-60MHz range, where it was
3.9dB out -- barely acceptable. You might want to ask
to see the site attenuation data from the labs
involved.

There can ! also be a big difference depending on the
measuring distance. Some standards allow extrapolation
to be done when measuring at some distance other than
the distance spelled out in the standard, but looking
at the difference in theoretical site attenuation
between a 3m and a 10m site will show that linear
extrapolation is not accurate through much of the
frequency range.

It is not unusual to get different results at the SAME
lab, depending on the person conducting the test. Like
it or not, some technicians are better or more
diligent when it comes to maximizing the emission by
playing with cable placement. When I worked at a
compliance test lab, I ticked off a few clients when
their product was passing by 2dB until I moved the
cables and rescanned only to fail them by 1dB.

Bob Richards
Square D.

--- Geetha Balasubramanian 
wrote:
> Dear Experts,
>
> One of our products had b! een tested by an
> accredited lab at US on May 2004 and found to be
> compliant with FCC, CE & VCCI. We started mass
> manufacturing the product and it is being sold in
> the US, Europe & Japan markets for the past 6
> months. Then came a requirement for that product in
> Taiwan market. Since the May 2004 testing at the US
> lab was not run in a BSMI approved system it became
> necessary for us to retest the same product. This
> time the product has been tested in an accredited
> lab in Taiwan. The Taiwan lab tests are showing that
> the product is not compliant to the CISPR22 and need
> a ferrite core in the USB cable to make it compliant
> to the standards. Product tested by the US lab &
> tested by he Taiwan lab are absolutely identical.
> As an engineer responsibled for compliancy I
> am a bit confused now. How to resolve this issue? I
> request you experts to g! uide me.
>
> Sincerely
>
> Geetha Balasubramanian
>


This message is from the IEEE Product Safety Engineering Society
emc-pstc discussion list. Website: http://www.ieee-pses.org/

To post a message to the list, send your e-mail to [email protected]

Instructions: http://listserv.ieee.org/listserv/request/user-guide.html

List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:

Scott Douglas [email protected]
Mike Cantwell [email protected]

For policy questions, send mail to:

Richard Nute: [email protected]
Jim Bacher: [email protected]

All emc-pstc postings are archived and searchable on the web at:

http://www.ieeecommunities.org/emc-pstc


This message is from the IEEE Product Safety Engineering Society
emc-pstc discussion list. Website: http://www.ieee-pses.org/

To post a message to the list, send your e-mail to [email protected]

Instructions: http://listserv.ieee.org/listserv/request/user-guide.html

List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:

Scott Douglas [email protected]
Mike Cantwell [email protected]

For policy questions, send mail to:

Richard Nute: [email protected]
Jim Bacher: [email protected]

All emc-pstc postings are archived and searchable on the web at:

http://www.ieeecommunities.org/emc-pstc



  _____  

Do you Yahoo!?
Yahoo!  <http://us.rd.yahoo.com/mail_us
taglines/virus/*http://promotions.yahoo.com/new_mail/static/protection.html>
Mail - Helps protect you from nasty viruses.
---------------------------------------------------------------- This message
is from the IEEE Product Safety Engineering Society emc-pstc discussion list.
Website: http://www.ieee-pses.org/ 

To post a message to the list, send your e-mail to [email protected] 


Instructions: http://listserv.ieee.org/listserv/request/user-guide.html 


List rules: http://www.ieee-pses.org/listrules.html 


For help, send mail to the list administrators: 


Scott Douglas [email protected] Mike Cantwell [email protected] 


For policy questions, send mail to: 


Richard Nute: [email protected] Jim Bacher: [email protected] 


All emc-pstc postings are archived and searchable on the web at: 


http://www.ieeecommunities.org/emc-pstc

---------------------------------------------------------------- This message
is from the IEEE Product Safety Engineering Society emc-pstc discussion list.
Website: http://www.ieee-pses.org/ 

To post a message to the list, send your e-mail to [email protected] 


Instructions: http://listserv.ieee.org/listserv/request/user-guide.html 


List rules: http://www.ieee-pses.org/listrules.html 


For help, send mail to the list administrators: 


Scott Douglas [email protected] Mike Cantwell [email protected] 


For policy questions, send mail to: 


Richard Nute: [email protected] Jim Bacher: [email protected] 


All emc-pstc postings are archived and searchable on the web at: 


http://www.ieeecommunities.org/emc-pstc 


Reply via email to