In article <[email protected]>, Doug Beckwith <[email protected]> writes >I need some help. IEC60601-1-2 calls up CISPR 11. Does EN60601-1-2 call >up CISPR11 or EN55011? > >The problem is this: CISPR 11 limits are taken at 10m and it does not >allow extrapolation of the data for measurements taken at 3m, whereas >EN55011 does in 7.2.3 (Amendment 1). Could we justify this provide it >was adequately documented in the report?
EN 'A' always calls up EN 'B' rather than IEC 'B', if EN 'B' exists. There is a trap for the unwary here; the ENs are listed in Annex ZA of the EN that is doing the referring, and you don't see Annex ZA unless you buy the EN. The references are sometimes NOT identical, or even technically equivalent, to the IEC Normative References in clause 2 of the IEC standard (some standards don't conform to the rule that they should be in clause 2). For example, EN 'B' may have a different date >from the latest IEC 'B', and/or it may include Common Modifications and/or Special National Conditions and/or A Deviations, that are not in IEC 'B'. -- Regards, John Woodgate, OOO - Own Opinions Only. The good news is that nothing is compulsory. The bad news is that everything is prohibited. http://www.jmwa.demon.co.uk Also see http://www.isce.org.uk This message is from the IEEE Product Safety Engineering Society emc-pstc discussion list. Website: http://www.ieee-pses.org/ To post a message to the list, send your e-mail to [email protected] Instructions: http://listserv.ieee.org/listserv/request/user-guide.html List rules: http://www.ieee-pses.org/listrules.html For help, send mail to the list administrators: Scott Douglas [email protected] Mike Cantwell [email protected] For policy questions, send mail to: Richard Nute: [email protected] Jim Bacher: [email protected] All emc-pstc postings are archived and searchable on the web at: http://www.ieeecommunities.org/emc-pstc

