HI Brian,

thanks for taking the time to do a round up... Much appreciated, and very 
interesting

Derek Walton
L F Research
----- Original Message ----- 
From: "Kunde, Brian" <[email protected]>
To: <[email protected]>
Sent: Thursday, September 13, 2007 4:06 PM
Subject: RE: Trouble with Chamber Lights - FINAL


> Hello All.
>
> For those of you who have been following this thread, let me fill you in
> on what we learned. Our most recent problem was caused by a single bulb,
> second nearest our antenna.  So we changed out all our bulbs with GE
> MVR400/U bulbs and we are back in business. Thanks to Mr. Woodgate, two
> bright individuals from GE Lighting contacted me and schooled me on
> metal halide lights.
>
> As you may know, we were fighting intermittent noise from our metal
> halide lights in our 10 meter chamber. We troubleshot the problem to a
> single GE/Tungsram MH400/U/TU bulb. We speculated the problem was caused
> by a failure in the bimetal switch and starter within the bulb. I have
> been told by GE this is incorrect.
>
> Metal Halide lights function by generating a controlled continuous arc
> within the arc tube between two electrodes. This arcing makes noise.
> Period! Once the bulbs warm up the arc stabilizes and it generates less
> RF noise. But for reasons beyond anyone's control, the arc can
> intermittently go unstable for a few seconds causing increased noise
> (ever see the lights change color). This problem can vary from bulb to
> bulb and change during the bulbs life. Changing out the bulbs may help
> for awhile but we have to consider this the nature of this lighting
> technology.
>
> The chamber manufacturers mount these lights above the chamber in a box
> (high-hat) lined with ferrite tile to minimize the RF noise from the
> lights. Though this is a good approach and works well most of the time,
> when the bulb arc goes unstable it can still interfer with your testing.
>
>
> So, what do we do about it?
>
> 1. It helps to properly "burn in" new bulbs.  Turn them on and let them
> burn in for 100 hours, turning them off for only 1 hour per day. This
> helps the arc to stabilize.
>
> 2. If you see noise from the lights during EMC testing, you will have to
> test with the lights off.
>
> 3. If you can't handle the darkness, then you need to have a secondary
> light source (stumble light) that will generate enough light to get you
> by, such as incandescent or halogen. Keep in mind that both these light
> technologies require more frequent bulb changes which can be a pain.
>
> Option 1:  If you can't wait for standard Metal halide lights to recycle
> back on, you may consider changing, at least a couple fixtures to Pulse
> Start or Instant On lights which come back on a lot faster, even when
> hot.  These lights may introduce new noise problems but if you have
> already decided to turn them off during testing then what does it
> matter.
>
> Option 2:  We found that the lights nearest the antenna are most likely
> to cause noise that we can see while testing. So we are going to re-wire
> our lights so we can turn off all but the two farthest from the antenna
> (over our turntable). That way when we test we can turn off 6 of 8. If
> we have trouble with these two, then we only have two lights to
> troubleshoot instead of 8.
>
> I have a meeting with ETS-Lindgren in a couple weeks and the lighting
> issue is on our agenda. If anything new comes from it I will pass it on.
>
> Good luck,
> The Other Brian
>
>
>
> -----Original Message-----
> From: [email protected] [mailto:[email protected]] On Behalf Of Kunde,
> Brian
> Sent: Monday, September 10, 2007 8:32 AM
> To: [email protected]
> Subject: RE: Trouble with Chamber Lights
>
> Does anyone have experience with "Pulse Start" Metal Halide lights?  It
> appears as if they took the unreliable starter electrode, bi-metal
> switch and resistor out of the bulb and somehow use a high voltage pulse
> from the ballast to start the bulb. One major advantage is the bulbs hot
> re-start much faster (don't have to wait for the bulb to cool off before
> the light will re-start).
>
> Anyway, we are speculating the problem with EMI noise from standard
> metal halide lights are in the bulbs and more specifically, in the
> unreliable bi-metal switch which falsely fires the starter electrode
> causing EMI noise.  Pulse Start bulbs seem to not have those parts so
> maybe they are better. Or maybe they have there own problems.  Any
> information would be appreciated.
>
> The Other Brian.
>
>
>
> -----Original Message-----
> From: [email protected] [mailto:[email protected]] On Behalf Of Kunde,
> Brian
> Sent: Friday, September 07, 2007 9:40 AM
> To: Ray, Darryl P; [email protected]
> Subject: RE: Trouble with Chamber Lights
>
> Here is an update on our progress.
>
> We are six days into our investigation and it appears the problem is
> with the bulbs themselves. We thought we were using the GE MVR400 as
> many has recommended (read below) but still having trouble with them.
> The bulbs have a bi-metal switch inside which shorts out the sparker
> (starter) once the bulb gets hot. We are speculating that this switch is
> intermittent and will sometimes start up the sparker after the bulb has
> heated up and this is the source of the emissions.  Because this problem
> is so intermittent, it makes it very hard to troubleshoot. After the
> lights warm up it may take hours to see if the problem is still there or
> not.
>
> We did the process of elimination method but we think we may have more
> than one bulb causing the problem. So as we are changing out the bulbs
> and plugging the lights back in, one at a time, we have to wait several
> hours or more to see if we can continue to the next light.
>
> It is strange that after 11 years with no problems we all of a sudden
> have nothing but trouble with these lights. We noticed the bulbs we
> installed six months ago, though they came in a box that said "GE", the
> bulbs say "TUNGSRAM" on the side. We found that Tungsram is a Hungarian
> company that GE bought. I wonder if Tungsram is making these bulbs now
> for GE and there is serious quality issues that is causing our problem.
> The new bulbs we are installing do say "GE" on them but they look
> identical to the Tungsram bulbs.
>
> Does anyone have good experience with other brands of metal halide bulbs
> like Sylvania? I can't believe a stupid light bulb has my chamber shut
> down for 6 days.
>
> Thanks to everyone for your replies. They have been most helpful. I keep
> posting on this topic so hopefully the information will be helpful to
> others someday.
>
> The Other Brian.
>
>
>
> -----Original Message-----
> From: Ray, Darryl P [mailto:[email protected]]
> Sent: Wednesday, September 05, 2007 7:24 PM
> To: Powell, Doug; Kunde, Brian; Stone, Richard; [email protected]
> Subject: RE: Trouble with Chamber Lights
>
>
> Brian,
>
> I have been involved with many EMC labs in particular the design and
> construction of semi anechoic chambers. This is  trickily problem as
> these lights are intended for warehouses not RF chambers. Metal Halide
> lights are very efficient but can be noisy.
>
> I have had lots of trouble with intermittent noise spikes from the Metal
> Halide lights. Based on what I have learned, the noise is probably
> coming from the lamp itself and not the ballast. I have had had pretty
> good luck with the GE MVR400 series (400 watt) of lamps.  Make sure you
> match the ANSI rating of the lamp to that of the ballast. This is
> critical.
>
> The bulbs do lose their lighting output rather quickly, and should be
> replaced about every 6 to 12 months. The lighting degradation is gradual
> and may not be noticed if your using the chamber frequently.  I have
> made measurements over time on a given changer and the reduction is
> indeed measurable.
>
> There always a chance of getting a noisy lamp so have extras on hand.
>
> Also the lights need to be on for at least 10 minutes otherwise they
> will be spikee.
>
> Hope this helps.
>
>
> Darryl Ray
>
>
> Darryl P. Ray
>
> Principal EMC Engineer
> Covidien - Respiratory Division
> 5823 Newton Drive
> Suite 110
> Carlsbad, CA  92008
>
> Voice: 760-603-5882
> Fax: 760-603-5369
> email: [email protected]
> Please note new company name (formerly Tyco Healthcare) address, phone
> number and email address
>
>
>
>
>
>
>
> -----Original Message-----
> From: [email protected] [mailto:[email protected]] On Behalf Of Powell,
> Doug
> Sent: Friday, August 31, 2007 2:54 PM
> To: Kunde, Brian; Stone, Richard; [email protected]
> Subject: RE: Trouble with Chamber Lights
>
> Brian,
>
> We have been using tungsten for several years without noise problems.
> However, the life of the bulbs is very short (2-3 months).  To combat
> this we purchase traffic signal bulbs.  There are several features in
> these bulbs for temperature extremes, rough handling, etc.  From my
> experience with these I think they will last about 2 years.
>
> The problem is that I cannot locate a flood light version of a traffic
> signal bulb, so this only works for general lighting.
>
>
>
> Thanks,
>
> -doug
>
> Douglas E. Powell
> Senior Compliance Engineer
> Corporate Product Compliance
>
> Advanced Energy Industries, Inc.
> Fort Collins, Colorado USA 80525
>
>
>
>
> -----Original Message-----
> From: [email protected] [mailto:[email protected]] On Behalf Of Kunde,
> Brian
> Sent: Friday, August 31, 2007 9:17 AM
> To: Stone, Richard; [email protected]
> Subject: RE: Trouble with Chamber Lights
>
> I forgot to add that we are seeing these spikes with nothing in the
> chamber (No EUT being tested) We have all other power going into the
> chamber disconnected, so the mast, turntable, applicance, camera and
> test power is all disconnected. We have evern disconnected all wiring
> going in and out of the chamber except for the Coax that goes to the
> receiver.
>
> The Other Brian
>
>
> -----Original Message-----
> From: Stone, Richard [mailto:[email protected]]
> Sent: Friday, August 31, 2007 11:11 AM
> To: Kunde, Brian
> Subject: RE: Trouble with Chamber Lights
>
> hi Brian,
> i find it hard to believe its the buld itself, the fixture and/or its
> wiring could be recieving transient or radiated emi within the ground
> plane from other equipment running locally. I dont recall the freq
> spikes that were showing up.
> I use standard 100wt flood lights in my chamber, the electrical to them
> is not the same as the electrical wiring for the turntable motor...
> i would check that as well, you can filter that to. but it means taking
> off the heavy turntable and doign some maitnenance,but it did work for
> me a while back.
> good luck
> Richard,
>
> -----Original Message-----
> From: Kunde, Brian [mailto:[email protected]]
> Sent: Friday, August 31, 2007 11:00 AM
> To: Stone, Richard; [email protected]
> Subject: RE: Trouble with Chamber Lights
>
>
> We have confirmed that the problem goes away when the lights are off.
> For instance, last night we ran our receiver in Max Hold with the lights
> off and not a single spike (16 hours). There are two 100db line filters
> on the power to the lights mounded on the highhat bulkhead. In one
> experiment, we bypassed the line filter for two reasons; If something in
> the filter was causing the spikes, then the problem should have gone
> away. It didn't. Also, if the source of the spikes were outside the
> chamber or on the power line feeding the filter/lights, then the spikes
> should have increased in amplitude. They didn't. So we feel the problem
> is within the chamber, which narrows it down to the lights themselves or
> the wiring.
>
> Today we will continue to try and narrow the problem down to a single
> fixture, but as some have suggested, the problem maybe an inherent
> problem with the ballasts or the bulbs. So the problem may be with all
> the lights. Any additional information on known good ballast or bulbs
> would be helpful. Our electrician said that he installed GE bulbs, but
> could also get Phillips bulbs if we thought that would make a
> difference.
>
> If the problem continues to allude us, we could become desperate enough
> to replace all eight fixtures with new ones. Any recommendations on that
> front? Other than LED, is there a better choice than metal halide for
> chamber illumination?
>
> Thanks again for everyone's help.
>
> The other Brian
> (head still bleeding)
>
>
>
> -----Original Message-----
> From: Stone, Richard [mailto:[email protected]]
> Sent: Friday, August 31, 2007 9:40 AM
> To: Hall, Ken; Kunde, Brian
> Cc: Crumm, Al
> Subject: RE: Trouble with Chamber Lights
>
> Have you confirmed this with the power shut off to the bulbs? Perhaps
> its the circuit powering the bulbs instead. Also i have seen in my
> chamber sporadic emi spikes due to bouncing of all the unused shielded
> cables whos ends are sitting on the conductive turntable floor, and
> bouncing up and down, that was unique, but its gone once i insulated
> them from floor.
>
> Can anyone reccommend a feed thru filter where all the BNC, RJ45
> connectors can be plugged in one side and then filtered out the other.
> This would be to stop ambients since the hole is 3 in diameter and 4
> inches in lenght. i am gettign about 10 ambients, 6 radio, 1 aircraft
> and 2 local ambients.
> my eut and support cables only fill about 30% of the hole.
>
> thnaks
> richard,
>
>
> -----Original Message-----
> From: [email protected] [mailto:[email protected]] On Behalf Of Hall,
> Ken
> Sent: Friday, August 31, 2007 9:24 AM
> To: Kunde, Brian; EMC PSTC
> Cc: Crumm, Al
> Subject: RE: Trouble with Chamber Lights
>
> Hello,
>
> We had/have similar prolems and, I think, finally figured out it was the
> base to lamp connections arcing.
>
> Please le tme know it thishelps.
>
> Regards,
>
> Ken Hall
>
> -----Original Message-----
> From: [email protected] [mailto:[email protected]] On Behalf Of Kunde,
> Brian
> Sent: Friday, August 31, 2007 6:06 AM
> To: EMC PSTC
> Subject: Trouble with Chamber Lights
>
> We have a Lindgren FACT10 (used to be called a FACT5) 10 meter
> semi-anechoic chamber built 11 years ago. Recently we have been seen
> sporadic spike emissions that interferes with our peak scans. We have
> narrowed the problem down to the chamber lights, which are 8 metal
> halide 400 watt lamps.
>
> We have replaced all the bulbs and the ballast but the problem is still
> there. We have replaced the line filters used on the light circuit; also
> no change.
>
> It is a real difficult problem to troubleshoot because the emissions are
> so sporadic. We started disconnecting the lights one by one. With fewer
> lights the problem seems to diminished, but it is like all the light
> generate this noise from time to time.
>
> Other than the startup noise from these metal halide lamps, I thought
> they are suppose to be very quite and not generate such EMI noise. I was
> hoping some of you chamber owners out there may have some insight into
> our problem.  We are running out of ideas here.
>
> Could we have obtained bad or cheep bulbs? Is there a manufacturer of
> replacement bulbs that are better? Is there a newer light technology
> that is better than the 11 year old metal halide lamps?
>
> Any and all suggestions are welcome and appreciated. We are stratching
> our heads bloody trying to figure this out.
>
> Thank you all,
> The Other Brian
>
>
> LECO Corporation Notice:  This communication may contain confidential
> information intended for the named recipient(s) only. If you received
> this by mistake, please destroy it and notify us of the error.  Thank
> you.
>
> -
> ----------------------------------------------------------------
> This message is from the IEEE Product Safety Engineering Society
> emc-pstc discussion list.    Website:  http://www.ieee-pses.org/
>
> To post a message to the list, send your e-mail to [email protected]
>
> Instructions:  http://listserv.ieee.org/request/user-guide.html
>
> List rules: http://www.ieee-pses.org/listrules.html
>
> For help, send mail to the list administrators:
>
>     Scott Douglas           [email protected]
>     Mike Cantwell           [email protected]
>
> For policy questions, send mail to:
>
>     Jim Bacher:             [email protected]
>     David Heald:            [email protected]
>
> All emc-pstc postings are archived and searchable on the web at:
>
>    http://www.ieeecommunities.org/emc-pstc
>
> -
> ----------------------------------------------------------------
> This message is from the IEEE Product Safety Engineering Society
> emc-pstc discussion list.    Website:  http://www.ieee-pses.org/
>
> To post a message to the list, send your e-mail to [email protected]
>
> Instructions:  http://listserv.ieee.org/request/user-guide.html
>
> List rules: http://www.ieee-pses.org/listrules.html
>
> For help, send mail to the list administrators:
>
>     Scott Douglas           [email protected]
>     Mike Cantwell           [email protected]
>
> For policy questions, send mail to:
>
>     Jim Bacher:             [email protected]
>     David Heald:            [email protected]
>
> All emc-pstc postings are archived and searchable on the web at:
>
>    http://www.ieeecommunities.org/emc-pstc
>
> LECO Corporation Notice:  This communication may contain confidential
> information intended for the named recipient(s) only. If you received
> this by mistake, please destroy it and notify us of the error.  Thank
> you.
>
> LECO Corporation Notice:  This communication may contain confidential
> information intended for the named recipient(s) only. If you received
> this by mistake, please destroy it and notify us of the error.  Thank
> you.
>
> -
> ----------------------------------------------------------------
> This message is from the IEEE Product Safety Engineering Society
> emc-pstc discussion list.    Website:  http://www.ieee-pses.org/
>
> To post a message to the list, send your e-mail to [email protected]
>
> Instructions:  http://listserv.ieee.org/request/user-guide.html
>
> List rules: http://www.ieee-pses.org/listrules.html
>
> For help, send mail to the list administrators:
>
>     Scott Douglas           [email protected]
>     Mike Cantwell           [email protected]
>
> For policy questions, send mail to:
>
>     Jim Bacher:             [email protected]
>     David Heald:            [email protected]
>
> All emc-pstc postings are archived and searchable on the web at:
>
>    http://www.ieeecommunities.org/emc-pstc
> This message, including any attachments, may contain information that is
> confidential and proprietary information of Advanced Energy Industries,
> Inc.  The dissemination, distribution, use or copying of this message or
> any of its attachments is strictly prohibited without the express
> written consent of Advanced Energy Industries, Inc.
>
>
> -
> ----------------------------------------------------------------
> This message is from the IEEE Product Safety Engineering Society
> emc-pstc discussion list.    Website:  http://www.ieee-pses.org/
>
> To post a message to the list, send your e-mail to [email protected]
>
> Instructions:  http://listserv.ieee.org/request/user-guide.html
>
> List rules: http://www.ieee-pses.org/listrules.html
>
> For help, send mail to the list administrators:
>
>     Scott Douglas           [email protected]
>     Mike Cantwell           [email protected]
>
> For policy questions, send mail to:
>
>     Jim Bacher:             [email protected]
>     David Heald:            [email protected]
>
> All emc-pstc postings are archived and searchable on the web at:
>
>    http://www.ieeecommunities.org/emc-pstc
>
> LECO Corporation Notice:  This communication may contain confidential
> information intended for the named recipient(s) only. If you received
> this by mistake, please destroy it and notify us of the error.  Thank
> you.
>
> -
> ----------------------------------------------------------------
> This message is from the IEEE Product Safety Engineering Society
> emc-pstc discussion list.    Website:  http://www.ieee-pses.org/
>
> To post a message to the list, send your e-mail to [email protected]
>
> Instructions:  http://listserv.ieee.org/request/user-guide.html
>
> List rules: http://www.ieee-pses.org/listrules.html
>
> For help, send mail to the list administrators:
>
>     Scott Douglas           [email protected]
>     Mike Cantwell           [email protected]
>
> For policy questions, send mail to:
>
>     Jim Bacher:             [email protected]
>     David Heald:            [email protected]
>
> All emc-pstc postings are archived and searchable on the web at:
>
>    http://www.ieeecommunities.org/emc-pstc
>
> LECO Corporation Notice:  This communication may contain confidential
> information intended for the named recipient(s) only. If you received
> this by mistake, please destroy it and notify us of the error.  Thank
> you.
>
> -
> ----------------------------------------------------------------
> This message is from the IEEE Product Safety Engineering Society
> emc-pstc discussion list.    Website:  http://www.ieee-pses.org/
>
> To post a message to the list, send your e-mail to [email protected]
>
> Instructions:  http://listserv.ieee.org/request/user-guide.html
>
> List rules: http://www.ieee-pses.org/listrules.html
>
> For help, send mail to the list administrators:
>
>     Scott Douglas           [email protected]
>     Mike Cantwell           [email protected]
>
> For policy questions, send mail to:
>
>     Jim Bacher:             [email protected]
>     David Heald:            [email protected]
>
> All emc-pstc postings are archived and searchable on the web at:
>
>    http://www.ieeecommunities.org/emc-pstc
>
> LECO Corporation Notice:  This communication may contain confidential 
> information intended for the named recipient(s) only. If you received this 
> by mistake, please destroy it and notify us of the error.  Thank  you.
>
> -
> ----------------------------------------------------------------
> This message is from the IEEE Product Safety Engineering Society
> emc-pstc discussion list.    Website:  http://www.ieee-pses.org/
>
> To post a message to the list, send your e-mail to [email protected]
>
> Instructions:  http://listserv.ieee.org/request/user-guide.html
>
> List rules: http://www.ieee-pses.org/listrules.html
>
> For help, send mail to the list administrators:
>
>     Scott Douglas           [email protected]
>     Mike Cantwell           [email protected]
>
> For policy questions, send mail to:
>
>     Jim Bacher:             [email protected]
>     David Heald:            [email protected]
>
> All emc-pstc postings are archived and searchable on the web at:
>
>    http://www.ieeecommunities.org/emc-pstc
> 

-

This message is from the IEEE Product Safety Engineering Society
emc-pstc discussion list.    Website:  http://www.ieee-pses.org/

To post a message to the list, send your e-mail to [email protected]

Instructions:  http://listserv.ieee.org/request/user-guide.html

List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:

     Scott Douglas           [email protected]
     Mike Cantwell           [email protected]

For policy questions, send mail to:

     Jim Bacher:             [email protected]
     David Heald:            [email protected]

All emc-pstc postings are archived and searchable on the web at:

    http://www.ieeecommunities.org/emc-pstc

______________________________________________________________________
This email has been scanned by the MessageLabs Email Security System.
For more information please visit http://www.messagelabs.com/email 
______________________________________________________________________

Reply via email to