Regarding Case 1 (and any other where voltages do not exceed the limits
in 6.3.1/6.3.2), it may be simpler to apply the fault testing in 9.1a)2)
rather than apply Basic Insulation. Of course a wrinkle would be if your
unlimited low voltage circuit is accessible, requiring you to deal with
the fire hazard in any external equipment the circuit could be connected
to (and possibly needing additional fault testing). If you do go for
Basic Insulation, then it seems that table 5 should be used for
clearance per 6.7.3.1a), since none of the exceptions in 6.7.3.1b) seem
to apply. Of course that table starts out at 50V. I would use the 50V
value from column 3 - 0.53mm.

Case 2 - seems about the same as Case 1, though you haven't specified
the secondary voltage.

Case 3 - Regarding separation from the output of the inverter and its 5V
supply, I'm not sure that spacings are required here. Is the 5V
accessible? What happens if the output of the inverter is shorted to the
input - are hazardous voltages per 6.3.1 accessible for any length of
time? 

Regarding clearances from the output of the inverter to other circuits,
the same basic rules apply as for Cases 1 and 2 - table 5 would be the
correct choice unless any of the exceptions in 6.7.3.1b) apply. Assuming
the mains voltage to the transformer with the 5V secondary is 230V and
OV cat II, then column 3 of table 5 would be appropriate.

Scott Aldous
Compliance Engineer
Advanced Energy
Tel: 970-407-6872
Fax: 970-407-5872


From: emc-p...@ieee.org [mailto:emc-p...@ieee.org] On Behalf Of
iun...@servomex.com
Sent: Tuesday, November 13, 2007 3:49 AM
To: emc-p...@ieee.org
Subject: EN 61010-1 - Clearances

Dear All,

I have three questions regarding the calculation of clearances in
accordance with the requirements of EN 61010-1. Can anybody out there
help
please?

Case 1 - Mains input transformer (230 Vac rms, Over-voltage category 2)
supplies an ac to dc converter providing +15 Vdc output (not energy
limited). To generate an energy limited supply, an appropriately
approved
PCB (Pollution degree 1, CTA> 175) mounted current limiting device is
fitted in series with the dc supply. How do I determine the clearances
around the unlimited +15 Vdc supply and the current limiting device?
Which
of the tables included in EN 61010-1 are appropriate and why?

Case 2 - PCB mounted mains transformer (230 Vac rms, Over voltage
category
2) is protected by an appropriately approved PCB mounted current
limiting
device in series with the secondary windings. How do I determine the
clearance distances around the tracks from the secondary winding up to
the
current limiting device? Which of the tables included in EN 61010-1 are
appropriate and why?

Case 3 - I am looking at designing a push-pull inverter powered by a 5
Vdc
supply derived from a mains transformer. What are the EN 61010-1 rules
governing the clearances around the outputs from this inverter and the
separation of the 5 V dc input and ac output circuits?

Many thanks in anticipation of your help.

Ian Unwin


Servomex Group Limited, Jarvis Brook, Crowborough, East Sussex, TN6 3DU,
England.
e-mail: i...@servomex.com
Company Registered in England: No.2170458
VAT No.: GB 522 6077 63

Register now at http://www.servomex.com/zap.html for an opportunity to
win
a great prize in our quarterly draw.

This email has been scanned for all viruses by the MessageLabs and
contains
information from Servomex which may be privileged or confidential. The
information is intended to be for the use of the individual(s) or entity
named above. If you are not the intended recipient be aware that any
disclosure, copying, distribution or use of the contents of this
information is prohibited. If you have received this electronic message
in
error, please notify us immediately. Servomex may, as part of its normal
activities, monitor, edit or censor the content of any information and
software, transmitted through, or stored on, its facilities.

-

This message is from the IEEE Product Safety Engineering Society
emc-pstc discussion list.    Website:  http://www.ieee-pses.org/

To post a message to the list, send your e-mail to emc-p...@ieee.org

Instructions:  http://listserv.ieee.org/request/user-guide.html

List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:

     Scott Douglas           emcp...@ptcnh.net
     Mike Cantwell           mcantw...@ieee.org

For policy questions, send mail to:

     Jim Bacher:             j.bac...@ieee.org
     David Heald:            emc-p...@daveheald.com

All emc-pstc postings are archived and searchable on the web at:

    http://www.ieeecommunities.org/emc-pstc

This message, including any attachments, may contain 
information that is confidential and proprietary information 
of Advanced Energy Industries, Inc.  The dissemination,
distribution, use or copying of this message or any of its
attachments is strictly prohibited without the express 
written consent of Advanced Energy Industries, Inc.

-

This message is from the IEEE Product Safety Engineering Society
emc-pstc discussion list.    Website:  http://www.ieee-pses.org/

To post a message to the list, send your e-mail to emc-p...@ieee.org

Instructions:  http://listserv.ieee.org/request/user-guide.html

List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:

     Scott Douglas           emcp...@ptcnh.net
     Mike Cantwell           mcantw...@ieee.org

For policy questions, send mail to:

     Jim Bacher:             j.bac...@ieee.org
     David Heald:            emc-p...@daveheald.com

All emc-pstc postings are archived and searchable on the web at:

    http://www.ieeecommunities.org/emc-pstc

______________________________________________________________________
This email has been scanned by the MessageLabs Email Security System.
For more information please visit http://www.messagelabs.com/email 
______________________________________________________________________

Reply via email to