Hello, Gary;
 
EN 60601-1-2: 2007 requirements  for ME specify +/- 2, 4, and 6 kV for Contact
discharge, and +/- 2, 4, and 8 kV for Air discharge.
 
Note that any patient-coupled signals (to or from the patient) must be
simulated, but that doing so must not provide an intentional conductive or
capacitive path to ground.
 
The Artificial Hand circuit is not used for this test, but is required for
Conducted Immunity, and EFT/B Immunity.
 
 
 
Regards,
David Raynes
Senior EMC Technologist
National Technical Systems
5151 - 47th Street NE
Calgary, Alberta
Canada T3J 3R2
 
Ph: 1-403-568-6605
Fax: 1-403-568-6970
[email protected]
 
From: McInturff, Gary [mailto:[email protected]] 
Sent: 12-Oct-10 17:52
To: [email protected]
Subject: [PSES] EN60601 ESD limits - 6 contact and 8kV air?
 
Can’t find my standard
 
Gary McInturff
 
-
----------------------------------------------------------------
This message is from the IEEE Product Safety Engineering Society emc-pstc
discussion list. To post a message to the list, send your e-mail to
<[email protected]>

All emc-pstc postings are archived and searchable on the web at
http://www.ieeecommunities.org/emc-pstc
Graphics (in well-used formats), large files, etc. can be posted to that URL. 
Website: http://www.ieee-pses.org/
Instructions: http://listserv.ieee.org/request/user-guide.html
List rules: http://www.ieee-pses.org/listrules.html 
For help, send mail to the list administrators:
Scott Douglas <[email protected]>
Mike Cantwell <[email protected]> 
For policy questions, send mail to:
Jim Bacher <[email protected]>
David Heald <[email protected]> 
-
----------------------------------------------------------------
This message is from the IEEE Product Safety Engineering Society emc-pstc
discussion list. To post a message to the list, send your e-mail to
<[email protected]>

All emc-pstc postings are archived and searchable on the web at
http://www.ieeecommunities.org/emc-pstc
Graphics (in well-used formats), large files, etc. can be posted to that URL. 
Website: http://www.ieee-pses.org/
Instructions: http://listserv.ieee.org/request/user-guide.html
List rules: http://www.ieee-pses.org/listrules.html 
For help, send mail to the list administrators:
Scott Douglas <[email protected]>
Mike Cantwell <[email protected]> 
For policy questions, send mail to:
Jim Bacher <[email protected]>
David Heald <[email protected]> 

Reply via email to