It said multiple listing is a terminology used in UL world long time ago.  It
does not exist in TUV laboratories and they have other complicated name.  That
is just a declaration of identical product with just different name and/or
cosmetics from the client, not from the test laboratory.  As the multiple
listing report issued by the test laboratory mis-leads the commercial sector
to believe that the testing laboratory has verified the sample they received
to be identical.  The laboratory just collects the DoI from the client and
issues certificate or covering with new model number.  Please correct me if
the understanding is inaccurate.

 

Regards,

 

Scott 

 

From: [email protected] [mailto:[email protected]] On Behalf Of Grace Lin
Sent: Wednesday, September 10, 2008 6:35 PM
To: [email protected]
Subject: Re: Labeling Requirements per Industry Canada

 

Dear Dennis and Others,

 

Thank you very much for your reply online and offline.  I contacted Industry
Canada as suggested.  IC refers to RSP-100 for an answer.  I excerpt a section
below for your reference.

 

Best regards,

Grace

 

"

5.3 Multiple Listing 

 

Multiple listing of a certified model is required when a manufacturer or
distributor wishes to list under its name and unique model number, certified
equipment of an original equipment manufacturer (OEM). 

 

A model of equipment may be multiple-listed to other manufacturers or
distributors based upon the approval granted to the original applicant and
certificate holder. 

 

In order to obtain a multiple-listing certification, the following
documentation must be submitted to the Bureau: 

 

(a) the model number and certification number of the approved equipment; 

 

(b) a signed letter from the original applicant and certificate holder
authorizing the Department to use information on file to grant a
multiple-listing certification. The name/model number and certification number
of the radio equipment must be shown. The letter must also declare that the
model to be multiple-listed is identical in design and construction to the
originally approved model; 

 

(c) a letter, from the applicant, requesting the certification; 

 

(d) completed and signed original copies of Appendix A and Appendix B; 

 

(e) completed and signed copies of Appendix A and Appendix B of RSS-102 -
Radio Frequency Exposure Compliance of Radiocommunication Apparatus (All
Frequency Bands); and 

 

(f) a drawing, sample or illustration of the product label.


"

On 9/9/08, dward <[email protected]> wrote: 

HI Grace

While it may be a desire of a marketing company to use their name, they are
not the certificate holder and thus putting only the marketing companies name
on the device would be incorrect.  The standard is pretty clear on this issue.
 The "the applicant's name (i.e. manufacturer's name, trade name or brand
name), model number and certification number" are to be on the device label.
This however, does not mean that in addition to these the marketing companies
name cannot be on the device.  

 

It should be remembered that the intent of the IC requirements is to identify
the device and link it to the manufacturer/applicant and the model number and
IC number. 

 

If a more specific interpretation from the IC is desired you may also contact
[email protected]. 

 

Thanks  

 

Dennis Ward 
Director of Engineering 
American TCB 
Certification Resource for the Wireless Industry www.atcb.com
<http://www.atcb.com/>  
703-847-4700 fax 703-847-6888 
direct - 703-880-4841 

 

From: [email protected] [mailto:[email protected]] On Behalf Of Grace Lin
Sent: Tuesday, September 09, 2008 11:17 AM
To: [email protected]
Subject: Labeling Requirements per Industry Canada

 

Dear Members,

 

Could someone please advise if a marketing company is able to mark an Industry
Canada certified OEM product by using the manufacturer's (applicant's) IC
number and the marketing company's name and model number?

 

I excerpt Section 5.2 of RSS-Gen below for your convenience.

 

If yes for the above question, is there any procedure we or our OEM supplier
has to follow (permissive change, etc.) ?

 

Thank you and look forward to hear from you.

 

Best regards,

Grace Lin

 

"

All Category I radio equipment intended for use in Canada shall permanently
display on each transmitter, receiver, or inseparable combination thereof, the
applicant's name (i.e. manufacturer's name, trade name or brand name), model
number and certification number. This information shall be affixed in such a
manner as not to be removable except by destruction or defacement. The size of
the lettering shall be legible without the aid of magnification but is not
required to be larger than 8-point font size. If the device is too small to
meet this condition, the information can be included in the user manual upon
agreement with Industry Canada. 

The label for medical implants which are designed to be used within the human
body, can be placed on the package and user manual.

The certification number is made up of a Company Number (CN) assigned by the
Bureau followed by the Unique Product Number (UPN), assigned by the applicant.

The certification number shall appear as follows: 

"IC: XXXXXX-YYYYYYYYYYY" 

Where: 

*       "XXXXXX-YYYYYYYYYYY" is the certification number; 
*       "XXXXXX" is the Company Number (CN) assigned by Industry Canada, made 
of at
most 6 alphanumeric characters (A-Z, 0-9), including a letter at the end of
the CN to distinguish between different company addresses; 
*       "YYYYYYYYYYY" is the Unique Product Number (UPN) assigned by the 
applicant,
made of at most 11 alphanumeric characters (A-Z, 0-9); and 
*       the letters "IC" have no other meaning or purpose than to identify the
Industry Canada certification number. 

Permitted alphanumerical characters used in the CN and UPN are limited to
capital letters (A-Z) and digits (0-9). An example of the new format for a
company having a CN of "21A" and wishing to use a UPN of "WILAN3" would thus
be: IC: 21A-WILAN3. Each equipment model shall be explicitly identified. The
use of characters, such as #, / or -, in the certification number is not
allowed. The use of "wild card" characters in the model number (for the
purpose of identifying multiple models with one name) is not allowed.

Equipment that has received certification but is not labelled with the
applicant's name, model number and the certification number as outlined above
is not considered certified.

Category II equipment shall be labelled in accordance with the requirements of
RSS-310. 

"

- ---------------------------------------------------------------- This
message is from the IEEE Product Safety Engineering Society emc-pstc
discussion list. Website: http://www.ieee-pses.org/ 

To post a message to the list, send your e-mail to [email protected] 

Instructions: http://listserv.ieee.org/request/user-guide.html 

List rules: http://www.ieee-pses.org/listrules.html 

For help, send mail to the list administrators: 

Scott Douglas [email protected] Mike Cantwell [email protected] 

For policy questions, send mail to: 

Jim Bacher: [email protected] David Heald: [email protected] 

All emc-pstc postings are archived and searchable on the web at: 

http://www.ieeecommunities.org/emc-pstc 


- ---------------------------------------------------------------- This
message is from the IEEE Product Safety Engineering Society emc-pstc
discussion list. Website: http://www.ieee-pses.org/ 

To post a message to the list, send your e-mail to [email protected] 

Instructions: http://listserv.ieee.org/request/user-guide.html 

List rules: http://www.ieee-pses.org/listrules.html 

For help, send mail to the list administrators: 

Scott Douglas [email protected] Mike Cantwell [email protected] 

For policy questions, send mail to: 

Jim Bacher: [email protected] David Heald: [email protected] 

All emc-pstc postings are archived and searchable on the web at: 

http://www.ieeecommunities.org/emc-pstc 

- ---------------------------------------------------------------- This
message is from the IEEE Product Safety Engineering Society emc-pstc
discussion list. Website: http://www.ieee-pses.org/ 

To post a message to the list, send your e-mail to [email protected] 

Instructions: http://listserv.ieee.org/request/user-guide.html 

List rules: http://www.ieee-pses.org/listrules.html 

For help, send mail to the list administrators: 

Scott Douglas [email protected] Mike Cantwell [email protected] 

For policy questions, send mail to: 

Jim Bacher: [email protected] David Heald: [email protected] 

All emc-pstc postings are archived and searchable on the web at: 

http://www.ieeecommunities.org/emc-pstc 

Reply via email to