Dear experts, 

I've been looking at the differences between EN 55020:2007 and EN
61000-4-6:2009 with respect to conducted RF immunity and I was hoping I could
get your professional opinion as I'm a bit confused.  Apologies in advance for
the long winded email.

1) 

EN 61000-4-6 works on the principle of having one injection point (e.g. CDN,
EM clamp) and one termination point for the common mode currents.  As I
understand it, every port that has a cable and AE connected must be decoupled
either by a CDN or a separate decoupling device (like an EM clamp?) to provide
a high impedance path to the common mode current.

EN 55020 clause 5.7 "Measurement of immunity to induced voltages" doesn't
specify this "single current loop and everything else decoupled" approach in
it's testing because injection onto the audio lines is differential mode onto
the signal conductor itself and not common mode onto the ground.

My first problem comes when considering injection of the unwanted signal onto
the mains using the Mains Stop Filter (MSF). In this case the signal is
injected common mode but the return path common mode impedance from the EUT to
the generator is not defined (unlike the 150-ohms via a CDN in 61000-4-6). 
Our TVs used for monitoring often have a grounded mains connection and our
antenna input from our streams network has the outside of the co-axial cable
grounded.

By having these connections to grounded inputs / monitoring equipment, the CM
impedance to common mode signals on the mains earth is very low.  In my
opinion, these lines either need to pass through a CDN or other decoupling
network of some kind with the same impedance requirements as set down in
EN55020 Annex C.1 b) (RF chokes shall provide a sufficiently high RF impedance
with respect to 150-ohms) over the whole frequency range) or even better in
EN61000-4-6 clause 6.2.4

In our case most of our products are Class II (no mains earth terminal) so the
impedance via the antenna input / monitoring equipment does not directly
affect the mains earth.  But the chassis still remains connected to earth as
mentioned above which could affect the results.  Should the antenna and the
feed to the monitoring equipment be decoupled from CM currents?

EN 55020 does not seem to have much detail regarding this - what are your
thoughts on the matter? 


2) 

The same question regarding the antenna feed for EN55020 testing also applies
to EN 61000-4-6 testing in that since it has a grounded coax outer it should
have it's own CDN or decoupling network.  Note: monitoring equipment in this
case has it's own CDN so isn't a worry.

Would you agree? 


3) 

The coupling unit type MC (mains leads) (EN 55020 figure C.2) is subject to
the note in Annex C.1 which states that the impedance of the CM choke in the
coupling unit does not effectivley meet the exact requirements of the standard
in the range 0.15MHz to 1.5MHz.  I don't believe that the MSF in EN55020
Figure D.3 does either.

I think a coupling unit that meets the specification of EN 61000-4-6 Figure
D.2 (and by extension EN61000-4-6 clause 6.2.4) should be adeqaute to use in
place of the MSF / coupling unit type MC in EN55020 testing for injecting onto
and decoupling the mains input.

Would you agree? 



Thanks in advance 
James 

James Pawson 
Leading Hardware Engineer 
EchoStar Europe 






-

This message is from the IEEE Product Safety Engineering Society emc-pstc
discussion list. To post a message to the list, send your e-mail to
<[email protected]>

All emc-pstc postings are archived and searchable on the web at
http://www.ieeecommunities.org/emc-pstc
Graphics (in well-used formats), large files, etc. can be posted to that URL. 

Website: http://www.ieee-pses.org/
Instructions: http://listserv.ieee.org/request/user-guide.html
List rules: http://www.ieee-pses.org/listrules.html 

For help, send mail to the list administrators:
Scott Douglas <[email protected]>
Mike Cantwell <[email protected]> 

For policy questions, send mail to:
Jim Bacher <[email protected]>
David Heald <[email protected]> 


Reply via email to