This is, for me, good stuff. I adore the physics of measurements.

Mr. Dudek has provided a link to a good info source that I was remiss for
not including. On this general subject of test data acceptance, please
note that all such UL guides are at

<www.ul.com/global/eng/pages/offerings/services/programs/dap/tools/>

To re-state my original question - what does the CFR say about temperature
measurement in reference to the NRTL program ?

Brian 

 > -----Original Message-----
 > From: [email protected] [mailto:[email protected]]On Behalf Of Dudek,
 > John (Corcom)
 > Sent: Tuesday, June 16, 2009 12:25 PM
 > To: [email protected]
 > Subject: RE: T-Couple Calibration
 >  
 >  From UL's website, acceptance of thermocouple guidelines 
 > for DAP participants.
 > 
 > http://www.ul.com/global/eng/documents/offerings/services/pro
grams/dap/tools/Equipment_Thermocouple.pdf
 > 
 > Regards
 > John F. Dudek
 > Manager, Product Safety Engineering
 > Corcom Products, Tyco Electronics Corp.
 > Mundelein, Il. USA
 > 847-573-6534 telephone
 > 847-680-0340 fax direct to PC
 > Mailto:[email protected]
 >  
 > 
 > -----Original Message-----
 > From: [email protected] [mailto:[email protected]] On Behalf 
 > Of Brian O'Connell
 > Sent: Tuesday, June 16, 2009 12:27 PM
 > To: [email protected]
 > Subject: RE: T-Couple Calibration
 > 
 > I buy thermocouple wire on separate spools so that I can 
 > control the twisted pair, then cut the twisted pairs into 
 > 1.5m segments. Then I choose a pair from the first third and 
 > a pair from the last third of the spools to verify. No 
 > complaints from any auditors to date. For calibration, look 
 > at ASTM E207/E220/E563 - but there are easier, more 
 > reasonable ways to verify instruments and temperature sensors.
 > 
 > I have not seen any particular requirement in 29 CFR 1910.7 
 > for NRTL thermocouple-based measurements/calibration. I 
 > would be very interested in any additional information that 
 > could be provided for NRTL temperature measurement requirements.
 > 
 > Let us talk about this concept called 'temperature', because 
 > I have seen some non-credible temperature data from CBTLs and NRTLs.
 > 
 > Temperature measurements are recorded for a small surface 
 > area of a larger mass, for a single instance in time. 
 > Temperature is a scalar quantity. Do not think of 
 > temperature measurements as vector quantities, and do not 
 > consider a temperature measurement to represent a 'constant' 
 > characteristic.
 > 
 > The NIST polynomials' accuracy for the common stuff (J, K, 
 > T) has a theoretical yield of about than 0.1 degC error 
 > through the full scale. The NIST polynomial error can be 
 > improved an order of mag for a delimited temperature range.
 > 
 > The reported accuracy of some instruments that are used for 
 > typical engineering measurements are about the same is the 
 > resolution (about 0.1 degC), which is not practical, and 
 > maybe not possible. For any of these common t/c types, the 
 > voltage gradient across a thermocouple wire pair is on the 
 > order of 100s of microvolts or perhaps 10's of millivolts 
 > for most product safety Type Test measurements. So the 
 > sensitivity of the thermocouple and its variable lead 
 > resistance and t/c attachment thermal impedance and thermal 
 > shunting and non-isothermal routing of t/c leads, the 
 > accuracy of the instrument, and the ambient noise conditions 
 > all conspire to make the resultant temperature measurement 
 > uncertainty, at best, 2 degC.
 > 
 > And a lab that reports temperatures to 0.1 degC is delusional.
 > 
 > There are some research/academic labs that may be able to 
 > control a test so the measurement uncertainty is 1 degC or 
 > better - but the practice is not reasonable for the 
 > engineering measurements in a product safety lab.
 > 
 > Let the shouting begin.
 > 
 > Brian
 > 
 > -----Original Message-----
 > From: [email protected] [mailto:[email protected]]On Behalf 
 > Of American Idle
 > Sent: Tuesday, June 16, 2009 8:38 AM
 > To: [email protected]
 > Cc: [email protected]
 > Subject: Re: T-Couple Calibration
 > 
 > You could pull the whole spool off and make your 2nd 
 > junction from the last bit, then re-wrap the whole spool :-)
 > 
 > I talked to a UL team lead on their DAP/ISO 17025 program 
 > and he had the following comments;
 > 
 > -This requirement is based on a CTL decision
 > -You must validate one TC from the beginning of the spool 
 > and one from the end of the spool with an RTL Calibrator or 
 > water bath method
 > -You risk all your previous data if the last TC you make 
 > from the spool doesn't calibrate right
 > 
 > He also stated that this requirement may change in the 
 > future because it doesn't make a lot of sense (and suggested 
 > that I bring this particular issue up for discussion if I 
 > happened to know anyone who sits on the Standards Commitee!).
 > 
 > As another poster stated, you may be better off purchasing 
 > pre-made thermocouples.  The only risk there is if your 
 > manufacturer goes out of business, your calibration 
 > certificates may become invalid.
 > 
 > -Ken Arenella
 > 
 > On Tue, Jun 16, 2009 at 9:50 AM, 
 > <[email protected]> wrote:
 > 
 > While we are on the subject. Anyone have an inexpensive 
 > solution to the big NRTL's new
 > Calibrated Thermocouple requirement?
 > 
 > As I read the spec it requires calibrating the first and 
 > last T-Couple off of the spool
 > minimum. Takes me a couple years to use 500ft. So I would 
 > technically only need to Calibrate
 > one a year.
 > 
 > Thanks in advance
 > 
 > John Merrill
 > Principal Product Safety Engineer
 > Schneider Electric

-

This message is from the IEEE Product Safety Engineering Society emc-pstc
discussion list. To post a message to the list, send your e-mail to
<[email protected]>

All emc-pstc postings are archived and searchable on the web at:
http://www.ieeecommunities.org/emc-pstc
Graphics (in well-used formats), large files, etc. can be posted to that URL.

Website:  http://www.ieee-pses.org/
Instructions:  http://listserv.ieee.org/request/user-guide.html
List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:
Scott Douglas <[email protected]>
Mike Cantwell <[email protected]>

For policy questions, send mail to:
Jim Bacher:  <[email protected]>
David Heald: <[email protected]>

Reply via email to