I used femm 4.2 and Hershey MC models to predict results of SAR testing and got 
better than 5% accuracy when compared to measurements on beef fat slabs from a 
local butcher. Just remember PSpice, LTspice, and femm ALL use peak voltage so 
must use sqrt(2)* Vrms value, else be off by 2:1 in power calculations.  The 
wrong direction.

Interesting question, whether the SAR rms value is independent of frequency. 
The models suggest you can calculate each and simply add the power. I do not 
remember any non-linearity coefficients in them, simply ables of conductivity 
vs freq etc.  But considering the perversity of nature and the conductivity 
being frequency dependent, you may find you get more heat than expected.  


I believe this company sells SAR predicting software, or will do it for you: 
notes from telco with Chad:
10/7/08  888 773 6266 x285  Chad  10:20  use SAR Models from Hershey Medical 
Center they contain characteristics of tissue versus frequency  

Remcom          1-888-773-6266
315 South Allen Street, Suite 222
State College, PA 16801 
 Tel: 814-861-1299 
 1-888-773-6266 (1-888-7REMCOM) Toll-Free in U.S. and Canada 
 Fax: 814-861-1308 
 1-888-973-6266 (1-888-9REMCOM) Toll-Free in U.S. and Canada 
  http://www.remcom.com


Good luck
Regards,
Robert


--- [email protected] wrote:

From: Robert Heller <[email protected]>
To: [email protected]
Subject: [PSES] SAR - MPE
Date:         Mon, 9 Apr 2012 11:13:32 -0500

Anyone familiar with SAR and MPE testing? 

If equipment passes SAR can you assume that it passes MPE?

Can simultaneous transmissions at two different frequencies be tested or 
is it calculated? Is the heating effect additive?

Thanks,
Bob Heller
3M Company, 76-1-01
St. Paul, MN 55107-1208
Tel: 651-778-6336
Fax: 651-778-6252


-
----------------------------------------------------------------
This message is from the IEEE Product Safety Engineering Society emc-pstc 
discussion list. To post a message to the list, send your e-mail to 
<[email protected]>

All emc-pstc postings are archived and searchable on the web at:
http://www.ieee-pses.org/emc-pstc.html

Attachments are not permitted but the IEEE PSES Online Communities site at 
http://product-compliance.oc.ieee.org/ can be used for graphics (in well-used 
formats), large files, etc.

Website:  http://www.ieee-pses.org/
Instructions:  http://listserv.ieee.org/request/user-guide.html
List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:
Scott Douglas <[email protected]>
Mike Cantwell <[email protected]>

For policy questions, send mail to:
Jim Bacher:  <[email protected]>
David Heald: <[email protected]>

-
----------------------------------------------------------------
This message is from the IEEE Product Safety Engineering Society emc-pstc 
discussion list. To post a message to the list, send your e-mail to 
<[email protected]>

All emc-pstc postings are archived and searchable on the web at:
http://www.ieee-pses.org/emc-pstc.html

Attachments are not permitted but the IEEE PSES Online Communities site at 
http://product-compliance.oc.ieee.org/ can be used for graphics (in well-used 
formats), large files, etc.

Website:  http://www.ieee-pses.org/
Instructions:  http://listserv.ieee.org/request/user-guide.html
List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:
Scott Douglas <[email protected]>
Mike Cantwell <[email protected]>

For policy questions, send mail to:
Jim Bacher:  <[email protected]>
David Heald: <[email protected]>

Reply via email to