You did not mention the product , but I assume that

 

IEC 60601-2-47:2012 concerns the basic safety and essential performance
of AMBULATORY ELECTROCARDIOGRAPHIC SYSTEMS

 

so you probably considering such a device.

 

 

I believe that if it is required to test essential performance, and for
thorough

the specific type of equipment a patient signal is required,  a
simulator is what you need to fulfil due diligence.

Especially if the device software autonomously draws any pertinent
conclusions from the measured signals

and the signals are not for  visual monitoring only.

 

A classical cardiographic device is basically a DM oscilloscope with a
high (assisted) CM suppression:

any device that can generate a large CM  and extremely small DM voltage
simultaneously will do.

 

In no way the testing for the integrity of stored data can be the only
compliance criterion here.

 

That said and concluded: I am not fully sure what your device is up
to.......

 

Gert Gremmen

 

 

 

Van: Silvia Diaz Monnier [mailto:silvi...@inti.gob.ar] 
Verzonden: dinsdag 28 maart 2017 20:12
Aan: EMC-PSTC@LISTSERV.IEEE.ORG
Onderwerp: [PSES] Radiated RF electromagnetic fields immunity test on
ambulatory electrocardiographic system

 

Hi,

 

the Radiated RF electromagnetic fields immunity test according to IEC
60601-2-47:2012, 202.6.2.3, makes an addition to IEC 60601-1-2:2007.
This addition requires to check that there is no loss of any stored
data. But collateral IEC 60601-1-2:2007, on 6.2.1.10 also requires to
verify the essential performance is not affected by noise on a waveform
in which the noise could interfere the diagnosis, treatment or
monitoring.

 

Test setup of IEC 60601-2-47 for that test do not require to simulate
the patient signal. Is that correct? 

If so, why 2-47 makes an addition instead of a replacement.

If not, why the test setup do not require to use a patient signal
simulator to check essential performance as other particular standards
IEC 60601-2-25 or IEC 60601-2-27.

 

That is, taking into account both standards, is it neccesary to check
both essential performance and the no loss of storaged data? Or only the
no loss of any stored data.

 

Thanks for your help.

Best regards,

Silvia

 

 

-
----------------------------------------------------------------

This message is from the IEEE Product Safety Engineering Society
emc-pstc discussion list. To post a message to the list, send your
e-mail to <emc-p...@ieee.org>

All emc-pstc postings are archived and searchable on the web at:
http://www.ieee-pses.org/emc-pstc.html

Attachments are not permitted but the IEEE PSES Online Communities site
at http://product-compliance.oc.ieee.org/ can be used for graphics (in
well-used formats), large files, etc.

Website: http://www.ieee-pses.org/
Instructions: http://www.ieee-pses.org/list.html (including how to
unsubscribe) <http://www.ieee-pses.org/list.html> 
List rules: http://www.ieee-pses.org/listrules.html 

For help, send mail to the list administrators:
Scott Douglas <sdoug...@ieee.org>
Mike Cantwell <mcantw...@ieee.org> 

For policy questions, send mail to:
Jim Bacher <j.bac...@ieee.org>
David Heald <dhe...@gmail.com> 


-
----------------------------------------------------------------
This message is from the IEEE Product Safety Engineering Society emc-pstc 
discussion list. To post a message to the list, send your e-mail to 
<emc-p...@ieee.org>

All emc-pstc postings are archived and searchable on the web at:
http://www.ieee-pses.org/emc-pstc.html

Attachments are not permitted but the IEEE PSES Online Communities site at 
http://product-compliance.oc.ieee.org/ can be used for graphics (in well-used 
formats), large files, etc.

Website:  http://www.ieee-pses.org/
Instructions:  http://www.ieee-pses.org/list.html (including how to unsubscribe)
List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:
Scott Douglas <sdoug...@ieee.org>
Mike Cantwell <mcantw...@ieee.org>

For policy questions, send mail to:
Jim Bacher:  <j.bac...@ieee.org>
David Heald: <dhe...@gmail.com>

Reply via email to