Hello
IEC 60601-1-2 allows testing of conducted and radiated emissions to CISPR 11;
CISPR 14-11; CISPR 32.
Concerning the use of CISPR 32:
According to Clause 7.1.1 General:
"Unless otherwise specified herein, ME
EQUIPMENT and ME SYSTEMS shall comply with CISPR 11"
According to Clause 7.1.3 Multimedia equipment:
"Multimedia equipment connected to ME EQUIPMENT and ME SYSTEMS shall comply
with CISPR 32. If CISPR 32 class A equipment is supplied as part of an ME
SYSTEM, the ME SYSTEM shall be classified class A."
A number of modern medical equipment includes devices that are clearly
multimedia equipment.
Are there any published guidelines for indicating when testing medical
equipment that includes multimedia devices to CISPR 32 instead of CISPR 11?
Best Regards,
David Shidlowsky
Technical Reviewer
[cid:[email protected]]
Address 1 Bat-Sheva St. POB 6117, LOD7120101 Israel
Tel: 972-8-9186100 Ext. 213 Direct: 972-8-9186113 Fax: 972-8-9153101
Mail: [email protected]<mailto:[email protected]> Web
www.itlglobal.org<http://www.itlglobal.org/>
[cid:[email protected]]Waze ITL: https://waze.com/ul/hsv8vccn2j
-
----------------------------------------------------------------
This message is from the IEEE Product Safety Engineering Society emc-pstc
discussion list. To post a message to the list, send your e-mail to
<[email protected]>
All emc-pstc postings are archived and searchable on the web at:
http://www.ieee-pses.org/emc-pstc.html
Attachments are not permitted but the IEEE PSES Online Communities site at
http://product-compliance.oc.ieee.org/ can be used for graphics (in well-used
formats), large files, etc.
Website: http://www.ieee-pses.org/
Instructions: http://www.ieee-pses.org/list.html (including how to unsubscribe)
List rules: http://www.ieee-pses.org/listrules.html
For help, send mail to the list administrators:
Scott Douglas <[email protected]>
Mike Cantwell <[email protected]>
For policy questions, send mail to:
Jim Bacher: <[email protected]>
David Heald: <[email protected]>